« Reply #60 on: April 12, 2019, 12:05:58 AM »

Model Number 300-20
Event Date 01/01/2016
Event Type  Malfunction   
Event Description
Patient underwent explant due to pain as reported in mfr. Report # 1644487-2016-01153. An analysis was performed on the returned lead portions. The electrodes portion of the leads was not returned for analysis and therefore a complete evaluation could not be performed on the entire lead product. Abraded openings were found on the outer and inner silicone tubing. These abraded openings and the cut ends that were made during the explanted process most likely provided the leakage path for the dried remnants of what appeared to have once been body fluids inside the outer silicone tubing. What appeared to be white deposits were observed in various locations. Eds (energy dispersion spectroscopy ¿ provides chemical or element identity/composition analysis) was performed on the deposit observed on the marked connector boot and identified the deposit as containing silicon, phosphorus, sodium, magnesium, sulphur and calcium. Refer to attached eds sheet for additional information. With the exception of the inner tubing abraded opening, the condition of the returned lead portions is consistent with conditions that typically exist following an explant procedure. The setscrew marks found on the connector pins provide evidence that, at one point in time, a good mechanical and electrical connection was present. Continuity checks of the returned lead portions were performed, during the visual analysis, with no discontinuities identified. Based on the findings, there is evidence of an inner tubing abraded opening.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=5778892
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #61 on: April 17, 2019, 06:23:22 AM »

Model Number 302-20
Event Date 01/01/2016
Event Type  Malfunction   
Event Description
Clinic notes were received for patient's referral for surgery due to high impedance. Clinic notes state that the vns is currently turned off secondary to high impendence and lead fracture. Notes also describe receiving a warning message about the device being able to deliver a maximum of 2 ma output current previously. Device diagnostics showed high lead impedance. No known surgical interventions have occurred to date.
 
Event Description
The patient underwent a full revision surgery on (b)(6) 2016. The explanted lead and generator were received on 10/04/2016. Analysis is underway but has not been completed to date.
 
Event Description
Various electrical loads were attached to the pulse generator and results of diagnostic tests demonstrate that accurate resistance measurements were obtained in all instances. A comprehensive automated electrical evaluation showed that the pulse generator performed according to functional specifications. The battery, 3. 006 volts, shows an ifi=no condition. There were no performance or any other type of adverse conditions found with the pulse generator. An analysis was performed on the returned lead portions and the reported allegations of lead fracture and high impedance, which were confirmed. A large portion of the lead assembly (body) including the electrode section was not returned for analysis; therefore a complete evaluation could not be performed on the entire lead product. Analysis confirmed discontinuity of both positive and negative quadfilar coils in the body region of the returned lead portions. Abraded openings of both outer and inner tubing near the break areas were also observed. Flat spots and pitting were observed on the coil surface. Determination could not conclusively be made on the fracture mechanism. Scanning electron microscopy was performed and identified the area on three of the broken coil strands as having evidence of a stress induced fracture (fatigue appearance) with mechanical damage, which most likely completed the fracture on one of the broken coil strands. Low magnification sem analysis of the quadfilar coil shows characteristics typical of a lead discontinuity which may include: material fracture, rough or pitted surface, thinned material thickness, electro-etching or material dissolution. The abraded openings found on the outer silicone tubing and the cut ends that were made during the explanted process, most likely provided the leakage path for the dried remnants of what appeared to have once been body fluids inside the outer silicone tubing. What appeared to be white deposits were observed in various locations. Eds (energy dispersion spectroscopy ¿ provides chemical or element identity/composition analysis) was performed on the deposit observed on the outer silicone tubing and identified the deposit as containing silicon, phosphorus, sodium, magnesium and calcium. With the exception of the observed discontinuities the condition of the returned lead portions is consistent with conditions that typically exist following an explant procedure. No other obvious anomalies were noted. The setscrew marks found on the lead connector pin provide evidence that, at one point in time, a good mechanical and electrical connection was present. Continuity checks of the returned lead portions were performed, during the visual analysis, and no other discontinuities were identified.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=5954926
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #62 on: April 19, 2019, 12:40:49 AM »

Device Problem Fracture
Event Date 11/03/2016
Event Type  Malfunction   
Event Description
It was reported that the patient's device showed high impedance. The device was programmed off and patient was referred for lead revision and prophylactic generator replacement. Patient's mother reported that the patient's seizure activity has increased every day since device has not been working. No additional relevant information has been received to date. No known surgical interventions have occurred to date.
 
Event Description
Patient underwent full revision surgery. The surgeon believes the single incision technique used during the initial implant surgery may have contributed to the lead break. The explanted lead was received. Analysis is underway but has not been completed to date.
 
Event Description
A portion of the lead assembly including the electrodes was not returned for analysis; therefore a complete evaluation could not be performed on the entire lead product. What appeared to be white deposits were observed in various locations. Energy dispersion spectroscopy (provides chemical or element identity/composition analysis) was performed on the deposit observed on the outer silicone tubing and identified the deposit as containing silicon, phosphorus, sodium, magnesium and calcium. The condition of the returned lead portion is consistent with conditions that typically exist following an explant procedure. No obvious anomalies were noted. The setscrew marks found on the lead connector pin provide evidence that, at one point in time, a good mechanical and electrical connection was present. Continuity checks of the returned lead portion were performed, during the visual analysis, with no discontinuities identified. Based on the findings in the lab, there is no evidence to suggest discontinuities in the returned portion of the device.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=6130470
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #63 on: April 24, 2019, 12:18:52 AM »

Model Number 302-20
Device Problems Mechanical Problem; Appropriate Term/Code Not Available
Event Date 12/08/2016
Event Type  Malfunction   
Event Description
During a generator replacement surgery, the old generator was unable to be interrogated. When the lead was connected to the new generator, high impedance was observed. The company representative performed 3 system diagnostics in a row. Each time, the output current was low, current delivered was 0. 25ma, and lead impedance was high >=10000 ohms. The old lead pin was fully inserted into the m 103 prior to system diagnostics. The surgeon looked under magnification to confirm, then disconnected lead, inspected lead and insertion point and confirmed again. High impedance was still present after confirming complete insertion. A full revision surgery was performed as a result. With the new lead connected to the new generator, impedance was within normal limits. The explanted lead has not been received to date.
 
Event Description
The generator and lead were received. Analysis is underway but has not been completed yet.
 
Manufacturer Narrative
(b)(4).
 
Event Description
The reported ¿end of service and low battery¿ allegations on the generator were confirmed in the lab; an open can measurement of the battery voltage determined that the battery was depleted. The end of service condition was the result of normal, expected battery depletion based on the battery life calculation, the electrical test results and the bench evaluation. Analysis of the generator in the lab concluded that no abnormal performance or any other type of adverse condition was found. An analysis was performed on the returned lead portions and the reported allegations of lead fracture and high impedance were confirmed. A portion of the outer silicone tubing and a portion of the connector ring quadfilar coil were not returned for analysis; therefore a complete evaluation could not be performed on the entire lead product. During the visual analysis of the returned lead, the lead coils appeared to be broken in several areas. Scanning electron microscopy performed on the several coil breaks identified the areas as having evidence of a stress induced fracture (fatigue appearance) with mechanical damage, some with pitting and some without pitting, evidence of a stress induced fracture (rotational forces) which most likely completed the fracture on two of the broken coil strands. The area on one of the two remaining broken coil strands was identified as having extensive pitting which prevented identification of the coil fracture type. The area on the last broken coil strand was identified as having evidence of being worn to the point of fracture which prevented identification of the coil fracture type. It is believed that stimulation was present for a certain period of time as evidenced by the presence of metal pitting. During the visual analysis of the returned 15mm portion the (-) green quadfilar coil appeared to be broken approximately 3mm past the end of the cut / torn inner silicone tubing. Low magnification sem analysis of the quadfilar coil shows characteristics typical of a lead discontinuity which may include: material fracture, rough or pitted surface, thinned material thickness, electro-etching or material dissolution. The slice mark found on the outer silicone tubing and the cut ends that were made during the explanted process, most likely provided the leakage path for the dried remnants of what appeared to have once been body fluids found inside the outer and inner silicone tubing. What appeared to be white deposits were observed in various locations. Eds (energy dispersion spectroscopy ¿ provides chemical or element identity/composition analysis) was performed on the deposit observed on the inner silicone tubing and identified the deposit as containing silicon, phosphorus, sodium, magnesium and calcium. With the exception of the observed discontinuities the condition of the returned lead portions is consistent with conditions that typically exist following an explant procedure. No other obvious anomalies were noted. The setscrew marks found on the lead connector pin provide evidence that, at one point in time, a good mechanical and electrical connection was present. Continuity checks of the returned lead portions were performed, during the visual analysis, and no other discontinuities were identified.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=6179592
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #64 on: April 24, 2019, 12:19:53 AM »

Model Number 302-20
Device Problems Fracture; Mechanical Problem
Event Date 12/07/2016
Event Type  Malfunction   
Event Description
Patient was scheduled for a prophylactic generator replacement on (b)(6) 2016 due to ifi - yes. Prior to the surgery, high impedance noted in three times and twice again in the or. The generator was replaced first and the lead was connected to the new m 106 generator. High impedance persisted and so the lead was revised. Impedance was within normal limits after the lead revision. The explanted devices have not been received to date.
 
Event Description
The explanted lead and generator is believed to be discarded.
 
Event Description
The explanted generator and lead were received. Analysis is underway but has not been completed to date.
 
Manufacturer Narrative
(b)(4).
 
Event Description
Various electrical loads were attached to the pulse generator and results of diagnostic tests demonstrate that accurate resistance measurements were obtained in all instances. A comprehensive automated electrical evaluation showed that the pulse generator performed according to functional. There were no performance or any other type of adverse conditions found with the pulse generator. An analysis was performed on the returned lead portions and the reported allegation of fracture of lead(s) was not confirmed. Note that the electrodes were not returned for analysis; therefore a complete evaluation could not be performed on the entire lead product. The abraded openings found in the outer and inner silicone tubing, slice marks found on the outer silicone tubing and the cut ends that were made during the explanted process, most likely provided the leakage path for the dried remnants of what appeared to have once been body fluids inside the outer silicone tubing. White deposits were observed in various locations. Eds (energy dispersion spectroscopy ¿ provides chemical or element identity/composition analysis) was performed on the deposit observed on the outer silicone tubing and identified the deposit as containing silicon, phosphorus, sodium and calcium. The condition of the returned lead portions is consistent with conditions that typically exist following an explant procedure. No obvious anomalies were noted except the half set of set screw marks observed near the end of the connector pin indicating the lead had not been fully inserted into the cavity of the generator. The additional set screw marks found on the lead connector pin provide evidence that, at one point in time, a good mechanical and electrical connection was present. Continuity checks of the returned lead portions were performed, during the visual analysis, and no discontinuities were identified. Based on the findings in the lab, there is no evidence to suggest discontinuities in the returned portions of the device.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=6198234&pc=LYJ
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #65 on: May 09, 2019, 06:48:52 AM »

Model Number 302-20
Device Problems Fracture; Adverse Event Without Identified Device or Use Problem
Event Date 03/11/2017
Event Type  Malfunction   
Event Description
Patient reported that he is getting painful, jolting stimulation when he turns his head. The patient was seen by the neurologist, who stated that the patient was being referred for vns battery replacement due to battery near end of life. This was reported to be an intervention for the painful stimulation as the physician believes the vns nearing end of life is the cause of the pain. Information was later received that the device is at end of service and patient was referred for replacement. Patient visited the er for pain in the vns area. Clinic notes were received for the replacement referral, indicating that the magnet did not abort an seizure. The physician attributed this to the low battery of the generator. In clinic notes, neos - yes was marked. Patient underwent generator replacement surgery but the explanted generator will not be returned per the explant facility.
 
Event Description
After generator replacement, the patient reported that he was unable to feel both normal and magnet stimulation. Upon interrogation of the device a high impedance warning message was observed. Diagnostic testing was performed. Impedance was found to be ok, but on the higher end (5221 ohms) with low output status. The generator was only supplying 1. 625 ma when normal mode was programmed to 1. 75 ma. There appears to be an intermittent impedance issues. X-rays were performed and a very obvious lead fracture was observed. The patient did not remember any falls or trauma that could have caused it however he knew something was wrong because he could not longer feel magnet stimulation and he had an increase in seizures. Patient underwent lead revision and the surgery reportedly went well but took a lot of time to complete. The explanted lead has not been received to date.
 
Event Description
The explanted lead was returned to the manufacturer. Analysis is underway but has not been completed.
 
Event Description
Analysis of the lead confirmed discontinuities of both positive and negative quadfilar coils in the body region of the returned lead portions. Abraded openings of both outer and inner tubing near the break locations were also observed. Scanning electron microscopy was performed on quadfilar coil break and identified the area as having evidence of a stress induced fracture (fatigue appearance) with mechanical damage, residual material and evidence of a stress induced fracture (rotational forces) which most likely completed the fracture. The area on the remaining broken coil strands was identified as being mechanically damaged which prevented identification of the coil fracture type with fine pitting. Pitting was observed on the coil surface in some areas. During the visual analysis, the negative electrode quadfilar coil appeared to be broken approximately 1 mm from the end of the abraded open / cut / outer / inner silicone tubes and the positive white electrode quadfilar coil appeared to be broken at approximately 2 mm. Low magnification sem analysis of the quadfilar coil shows characteristics typical of a lead discontinuity which may include: material fracture, rough or pitted surface, thinned material thickness, electro-etching or material dissolution. The abraded openings found on the outer and inner silicone tubes, most likely provided the leakage path for the dried remnants of what appeared to have once been body fluids inside the outer and inner silicone tubes. What appeared to be white deposits were observed in various locations. Energy dispersion spectroscopy was performed on the deposit observed on the outer silicone tubing and identified the deposit as containing silicon, phosphorus, sodium, magnesium sulphur and calcium. With the exception of the abraded openings observed on the outer and inner silicone tubes and the observed discontinuities the condition of the returned lead portions is consistent with conditions that typically exist following an explant procedure. No other obvious anomalies were noted. The set screw marks found on the lead connector pin provide evidence that, at one point in time, a good mechanical and electrical connection was present. Continuity checks of the returned lead portions were performed, during the visual analysis, and no other discontinuities were identified.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=6509179
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #66 on: May 14, 2019, 12:12:32 AM »

Model Number 104
Device Problem Adverse Event Without Identified Device or Use Problem
Event Date 12/08/2016
Event Type  Injury   
Event Description
The nurse practioner is referring patient for full revision due to migration. Replacement was stated to be necessary because patient's current vns has migrated to the edge of the chest wall, rendering access very difficult. No known surgical interventions have occurred to date. Additional relevant information has not been received to date.
 
Event Description
Additional information was received that the device migration occurred between visits in (b)(6) 2016 and (b)(6) 2017. There are no known causes for the migration such as trauma or manipulation. Surgical intervention is planned to preclude a serious injury. The previous implant surgeon confirmed that a non absorbable suture was used to secure the generator to fascia during the implant. No known surgery has occurred to date.
 
Manufacturer Narrative
(b)(4).
 
Event Description
The explanted generator and lead were received. The pulse generator diagnostics were as expected for the programmed parameters. A comprehensive automated electrical evaluation showed that the pulse generator performed according to functional specifications. There are no performances or any other type of adverse conditions found with the pulse generator. Analysis of the lead is underway but has not been completed to date.
 
Event Description
Analysis of the received lead was completed. The electrodes were not returned for analysis and therefore a complete evaluation could not be performed on the entire lead product. The abraded openings found on the outer silicone tubing, most likely provided the leakage path for the dried remnants of what appeared to have once been body fluids inside the outer silicone tubing. What appeared to be white deposits were observed in various locations. Energy dispersion spectroscopy (provides chemical or element identity/composition analysis) was performed on the deposit observed on the unmarked connector boot and identified the deposit as containing silicon, phosphorus, sodium, magnesium and calcium. The condition of the returned lead portion is consistent with conditions that typically exist following an explant procedure. No obvious anomalies were noted. The set screw marks found on the lead connector pins provide evidence that, at one point in time, a good mechanical and electrical connection was present. Continuity checks of the returned lead portion were performed, during the visual analysis, with no discontinuities identified. Based on the findings in the lab, there is no evidence to suggest an anomaly with the returned portion of the device.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=6455535
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #67 on: June 15, 2019, 10:43:51 PM »

Model Number 302-20
Device Problems Fracture; Appropriate Term/Code Not Available
Event Date 01/01/2017
Event Type  Malfunction   
Manufacturer Narrative

Event Description
During a consult for generator replacement due to battery depletion, a lead fracture was observed in x-rays. The generator battery was reported to be dead. Both the generator and lead were replaced. The explanted lead was received. Analysis is underway but has not been completed to date.
 
Manufacturer Narrative
 
Event Description
During the visual analysis, multiple quadfilar coils appeared to be broken. Scanning electron microscopy was performed and identified the area as having evidence of a stress induced fracture. Determination could not conclusively be made on the fracture mechanism. Pitting and residual material were observed on the coil surface. Scanning electron microscopy identified an area on three of the broken coil strands as being mechanically damaged which prevented identification of the coil fracture type with pitting on one and residual material on the other two broken coil strands. It is believed that stimulation was present for a certain period of time as evidenced by the presence of metal pitting. Low magnification sem analysis of the quadfilar coil shows characteristics typical of a lead discontinuity which may include: material fracture, rough or pitted surface, thinned material thickness, electro-etching or material dissolution. The abraded opening / torn areas found on the outer silicone tubing, most likely provided the leakage path for the dried remnants of what appeared to have once been body fluids inside the outer silicone tubing. What appeared to be white deposits were observed in various locations. Eds (energy dispersion spectroscopy ¿ provides chemical or element identity/composition analysis) was performed on the deposit observed on the outer silicone tubing and identified the deposit as containing silicon, phosphorus, sodium, magnesium and calcium. Since a portion of the lead assembly including the electrode array section was not returned for analysis, an evaluation and resulting commentary cannot be made on that portion of the lead.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=7042802
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #68 on: June 17, 2019, 03:59:18 AM »

Model Number 302-20
Device Problems High impedance; Appropriate Term/Code Not Available
Event Date 12/08/2016
Event Type  Malfunction   
Manufacturer Narrative

Event Description
Patient was referred for a prophylactic generator replacement. During the replacement surgery, high impedance was observed with dcdc - 7 on system diagnostic test with the existing device. The surgeon inserted the lead pin into the generator a couple of times to rule out incomplete pin insertion but continued to get high impedance. The lead and generator were therefore replaced. The explanted devices have not been received to date.
 
Event Description
The explanted devices were received. Analysis is underway but has not been completed to date.
 
Event Description
The generator was explanted for prophylactic reasons. The device performed according to functional specifications of the current automated final test. Analysis of the generator in the analysis lab concluded that no abnormal performance or any other type of adverse condition was found. An analysis was performed on the returned lead portions and the reported allegations of fracture of lead were confirmed. The electrodes were not returned for analysis; therefore a complete evaluation could not be performed on the entire lead product. During the visual analysis of the returned portion what appeared to be pitting and rust-like deposits were observed on the connector pin surface. Energy dispersion spectroscopy (eds), which provides chemical or element identity/composition analysis, was performed on the connector pin rust-like deposits and identified the area as consisting of sodium, silicone, phosphorus, nickel, chromium and iron. Another eds procedure was performed on the clean connector pin surface and identified the area as consisting of chromium, silicone, sulphur, iron and nickel. A definite cause for the pitting could not be determined based on the lead portions returned. Several broken coils were observed. Scanning electron microscopy (sem) was performed and identified the some areas as being mechanically damaged (smooth surfaces) which prevented identification of the coil fracture type with pitting on the broken coil strands. It is believed that stimulation was present for a certain period of time as evidenced by the presence of metal pitting. Low magnification sem analysis of the quadfilar coil shows characteristics typical of a lead discontinuity which may include: material fracture, rough or pitted surface, thinned material thickness, electro-etching or material dissolution. No other obvious anomalies were noted. The set screw marks found on the lead connector pin provide evidence that, at one point in time, a good mechanical and electrical connection was present.
 
Event Description
It was noted during a review of a programming data card that high impedance was observed during system diagnostics the patient's generator implant surgery on (b(6) 2016. This is the generator that was explanted with the lead. It is unclear if this is related to pin insertion troubleshooting event or related to the lead fracture. No further relevant information has been received to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=7151020
Logged