« Reply #90 on: January 14, 2019, 04:33:12 AM »

Model Number 302-20
Event Date 04/10/2014
Event Type  Malfunction   
Manufacturer Narrative
Manufacturing device history records were reviewed. Review of the lead device history records confirmed all quality tests were passed prior to distribution.
 
Manufacturer Narrative
Review of the available programming and diagnostic history. Device failure is suspected, but did not cause or contribute to a death or serious injury.
 
Event Description
An implant card was received indicating that the vns patient was experiencing spasms and pain at the generator and lead sites prior to generator replacement surgery on (b)(6) 2014. The generator was replaced due to battery depletion.
 
Event Description
It was reported that the patient has had multiple issues including pain and decreased appetite since the lead was repaired with silicone. The pain in the neck area was intense, and the patient kept ice applied and took medication as a result. The patient had lead replacement surgery on (b)(6) 2014. The patient reported on (b)(6) 2014 improved adverse symptoms and feeling much better since the surgery. The explanted lead is not being returned to the manufacturer for analysis.
 
Event Description
It was reported that during generator replacement surgery on (b)(6) 2014, the surgeon repaired two large tears in the lead with silicone. Since the surgery, the patient was having difficulty turning her neck. When she turned her neck, the patient became nauseous from pain in her throat. The patient was unable to eat due to the nausea and had lost weight. The patient was also experiencing tinnitus with her head turned and having daily headaches. The patient¿s device was recently programmed on. The patient previously had a surgical consult on (b)(6) 2013 for a possible lead migration. The surgeon determined there was no lead migration and stated that the patient¿s device was working well. Diagnostic results showed normal device function as of (b)(6) 2013. Review of the available programming and diagnostic history showed normal diagnostic results through 11/23/2011. No known interventions have occurred to date. Attempts for additional relevant information have been unsuccessful to date.
 
Event Description
It was reported that the patient was going to be evaluated by another surgeon for a second opinion. The patient was referred for generator and lead replacement surgery. No known surgical interventions have occurred to date.
 
Event Description
The explanted generator is not available to return to the manufacturer for analysis.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3838722
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 57640


« Reply #91 on: January 14, 2019, 04:34:30 AM »

Model Number 302-20
Event Date 04/13/2014
Event Type  Injury   
Event Description
It was reported on the implant card that the reason for lead replacement was a ¿lead discontinuity¿. It was reported that the hospital discards the explants and therefore cannot be returned for product analysis.
 
Event Description
Clinic notes dated (b)(6) 2014 note that the patient was still having symptoms after generator settings changes. It was noted that the patient needs replacement of the lead to see if that will help with her symptoms and allow her to receive full vns therapy. It was noted that no other major changes have occurred in the patient's medical history. An implant card was received indicating that the patient underwent lead replacement on (b)(6) 2014. The explanted device has not been received for analysis to date.
 
Manufacturer Narrative
Describe event or problem; corrected data: inadvertently did not include that the lead replacement was due to a lead discontinuity on follow-up report #1.
 
Manufacturer Narrative
Review of the available programming and diagnostic history.
 
Event Description
It was reported that the vns patient went to the er on (b)(6) 2014 due to painful erratic stimulation. The er physician and neurologist stated that the patient¿s generator may be at end of service. The patient had not felt stimulation for the past eight months. Before going to the er, the patient was on a swing and was hanging by her left arm. The patient began experiencing painful stimulation at 45 second intervals. The patient was also experiencing tightening in her chest and reported that her magnet was not working. The patient had to press the magnet extremely hard against her generator to disable her device. Taping the magnet over the generator was ineffective in disabling the device. The patient was uncertain if her generator had migrated. The patient was seen by her neurologist on (b)(6) 2014. The neurologist decreased the patient¿s device settings during the office visit. The neurologist noted that the patient¿s voice alteration had significantly increased with stimulation. The patient felt that stimulation was traveling up her ear and around her temple. The physician stated that the generator pocket may have expanded and the generator may have migrated causing the magnet to activate magnet mode stimulation instead of disabling the device. The patient¿s device was tested during the office visit and diagnostic results showed lead impedance within normal limits. Further follow-up revealed that the patient underwent generator replacement surgery on (b)(6) 2014 but continued to have painful stimulation after surgery. The replacement device had been programmed on to the patient¿s previous device settings. A radiology report was received but did not observe any issues with the generator and lead. The patient was referred for lead replacement surgery. Lead replacement surgery has not occurred to date. Review of the available programming and diagnostic history showed normal diagnostic results through (b)(6) 2012. A battery life calculation using the available programming history showed approximately 1. 77 years until eri = yes. The explanting facility will not return explanted devices to the manufacturer for analysis; therefore, no analysis can be performed.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3840370
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 57640


« Reply #92 on: January 20, 2019, 03:20:39 AM »

Model Number 302-20
Event Date 05/13/2014
Event Type  Injury   
Manufacturer Narrative

Event Description
It was reported that the vns patient was referred for surgery to explant his device due to discomfort and pain from a lead pulling sensation whenever he turned his head. The patient¿s device was disabled in 2011 due to lack of efficacy in treating the patient¿s depression. It was later determined that the patient did not have depression but was suffering from an anxiety disorder. The patient¿s generator and lead were explanted on (b)(6) 2014. The explanted generator and lead have been returned to the manufacturer where analysis is currently underway.
 
Event Description
Additional information was received stating that the vns patient underwent surgery for patient comfort and not to preclude a serious injury. The patient was a non-responder to vns therapy; therefore, the patient¿s generator was explanted and not replaced. Analysis of the returned generator and lead was completed. Monitoring of the generator output signal showed no signs of variation in the pulse generator¿s output signal and demonstrated that the device provided the expected level of output current. The pulse generator diagnostics were as expected for the programmed parameters. The device performed according to functional specifications. Analysis concluded proper functionality of the pulse generator and that no abnormal performance or any other type of adverse condition was found. The condition of the returned lead portion is consistent with conditions that typically exist following an explant procedure. No obvious anomalies were noted except for the setscrew marks observed near the end of the connector pin, indicating the connector pin had not been fully inserted into the cavity of the pulse generator at one time. The connector pin was inserted into the cavity of a test generator to verify a dimensional issue in the connector portion of the lead wasn¿t preventing the connector pin from being fully inserted into a generator. No obstructions were noted. The marks are evidence of a potentially insufficient mechanical contact between conductive surfaces of the generator and connector ring, resulting in a suspect electrical connection to the lead. However, based on the location of the setscrew marks on the pin, it is unknown whether a good electrical connection was present for the connector ring. Canted spring marks were not observed on the rear end of the small o-ring. Continuity checks of the returned lead portion were performed, during the visual analysis, and no discontinuities were identified. Based on the findings, there is no evidence to suggest an anomaly with the returned portions of the device which may have contributed to the stated complaints. Note that since a portion of the lead assembly including the electrode array section was not returned for analysis, an evaluation and resulting commentary cannot be made on that portion of the lead.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3854239
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 57640


« Reply #93 on: January 20, 2019, 03:21:40 AM »

Model Number 106
Device Problem Adverse Event Without Identified Device or Use Problem
Event Date 01/26/2018
Event Type  Injury   
Manufacturer Narrative
(b)(4).
 
Event Description
Further information was received that despite the generator being disabled, the patient continued to experience the previously reported adverse events, specifically pain. It was stated that the paint occurred in the generator area and was associated with left upper extremity pain and some numbness. Because of this, the patient requested both the lead and generator be removed. The physician reportedly advised that the pain the patient was feeling was likely not from vns, but a definitive assessment was not made. No surgical intervention has occurred to date. No additional relevant information has been received.
 
Event Description
Report received that a pain had pain that radiated from her generator site to her neck, face, and right side of her head. This pain was reportedly worse upon exertion but also occurred while sitting. The pain had reportedly occurred for three weeks with a few severe episodes. The patient also reported feeling fatigued. The patient reportedly went to the hospital during one severe painful experience where cardiac work was done. Although the results were negative, there was reportedly some bradycardia. The physician reported that the vns was checked and the settings were "low". The impedance was also reportedly checked and was found to be within normal limits. The physician elected to disable the device but indicated he would refer the patient for follow-up with a surgeon for reevaluation of other etiologies. No additional relevant information has been received to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=7352997
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 57640


« Reply #94 on: February 12, 2019, 03:23:25 AM »

Event Date 01/26/2015
Event Type  Malfunction   
Event Description
Additional information was received stating that the vns patient underwent generator and lead replacement surgery on (b)(6) 2015. The explanting facility will not return explanted devices to the manufacturer for analysis; therefore, no analysis can be performed. Clinic notes were received indicating that the vns patient presented with new symptoms during an office (b)(6) 2015. The right side of the patient¿s lower face and upper neck was twitching which would occur for a few minutes and then stop. The patient also was having sharp pains on the left side of her head that occurred daily. High impedance was observed during the office visit and the patient¿s device was not disabled.
 
Event Description
It was reported that the vns patient¿s device was tested during an office visit on (b)(6) 2015 and system diagnostic results revealed high impedance (dcdc ¿ 7). The patient¿s device showed lead impedance within normal limits (dcdc ¿ 2) at the previous office visit on (b)(6) 2014 the patient was referred for surgery but no known surgical interventions have occurred to date.
 
Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4532847
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 57640


« Reply #95 on: February 13, 2019, 03:02:24 AM »

Model Number 304-20
Event Date 01/03/2015
Event Type  Malfunction   
Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to a death or serious injury.
 
Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to a death.
 
Event Description
Additional information was received that the patient is undergoing ent procedures to treat painful stimulation in the neck and head, voice alteration, and coughing that he has been experiencing since early january. However, the device being disabled has also helped diminish these symptoms as they were thought to be related to stimulation during high impedance. No known vns surgical interventions have occurred to date.
 
Event Description
Additional information was received that the patient underwent a recent brain surgery and the vns is still disabled. The patient's previous neurologist has no additional information to provide. No known surgical interventions have occurred to date.
 
Event Description
It was reported that device diagnostics resulted in high impedance (6592 ohms). Previous device diagnostics on (b)(6) 2015 were within normal limits (2736 ohms). X-rays were taken and did not identify any obvious discontinuities with the vns system. The generator was programmed off and the patient was referred to surgeon. No known surgical interventions have occurred to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4516986
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 57640


« Reply #96 on: February 17, 2019, 06:22:08 AM »

Model Number 302-30
Event Date 01/01/2015
Event Type  Injury   
Event Description
Analysis of the generator was completed on 03/05/2015. A comprehensive automated electrical evaluation showed that the pulse generator performed according to functional specifications.
 
Manufacturer Narrative

Event Description
Initially it was reported that the patient was scheduled for generator and lead replacement for an unknown reason. It was reported that the generator was replaced due to end of service and that it was believed that the patient experienced an increase in seizures, but the pre-vns baseline frequency is unknown. The patient requested to have the lead replaced due to the lead having become more prominent over the last year and pain in the neck when turning head in different directions for 2 weeks prior to surgery. The reason for the protrusion is unknown. Device diagnostics were within normal limits. It was reported that the lead was replaced to preclude a serious injury and for patient comfort. The pain and protrusion resolved following generator and lead replacement surgery. The explanted lead was discarded during surgery. The explanted generator was returned for analysis on 02/10/2015. Analysis is underway, but has not been completed to date. No additional relevant information has been received to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4570551