« on: January 28, 2017, 03:21:57 AM »

Model Number MODEL 250
Event Date 01/01/2010
Event Type Malfunction
Event Description
It was reported that the handheld's ac adapter cable had a defect causing the handheld to not charge. Troubleshooting was performed, but the problem persisted. Product was returned to the manufacturer, but analysis is pending.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1948993
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #1 on: January 28, 2017, 03:23:10 AM »

Model Number MODEL 250
Event Date 12/01/2010
Event Type Malfunction
Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to death or serious injury.

Event Description
It was reported that the manufacturer representative's handheld battery was not charging and that the ac adapter cord had to be held in a specific position for the device to charge. The handheld computer, software flashcard, and the ac adapter cord are en route to the manufacturer headquarters where analysis will be performed.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1984755
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #2 on: January 28, 2017, 03:24:15 AM »

Model Number MODEL 250
Event Date 01/01/2010
Event Type Malfunction
Event Description
It was reported that the physician's hand held was charging intermittently and it is suspected that the ac adapter cord is not working properly. The physician requested a new hand held to replace the non-working device. Good faith attempts to obtain the non-working device. Good faith attempts to obtain the non-working device for analysis have been made, but the device has not been returned to date.

Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1621078
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #3 on: January 28, 2017, 03:25:20 AM »

Model Number 250
Event Date 02/01/2010
Event Type Malfunction
Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to death or serious injury.

Event Description
It was reported that the physician's hand held was not charging properly with the ac adapter cable. The physician used another working ac adapter cable and the hand held was able to be charged with no further problems. Good faith attempts to obtain power cord (ac adapter cable) for analysis have been made, but the cord has not been returned to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1639311
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #4 on: January 28, 2017, 03:26:55 AM »

Model Number MODEL 250
Event Date 06/01/2010
Event Type Malfunction
Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to a serious injury or death.

Event Description
It was reported to manufacturer that the distributor in (b) (4) found a bad electrical contact of the cable that connected the ac adapter with the handheld computer. The product has not been returned to manufacturer for analysis. Good faith attempts to obtain additional information have been unsuccessful to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1773954
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #5 on: January 28, 2017, 03:27:52 AM »

Model Number 250
Event Date 06/21/2010
Event Type Malfunction
Event Description
It was reported that the physician's hp hand held would not hold a charge any longer. This event began recently, and it was noted that the product is stored in the office and is always plugged in while not in use. The physician requested a new hand held to replace the non-working device. The hand held and software flashcard were returned to manufacturer for analysis where there were no anomalies found and both the software and hand held performed according to specifications; however, the ac adapter was not returned for analysis and therefore, the ac adapter is not able to be ruled out as the cause of the issue the site was experiencing. Good faith attempts are underway to obtain the ac adapter cable from the site for analysis.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1786065
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #6 on: January 28, 2017, 03:29:19 AM »

Model Number 250
Event Date 04/01/2010
Event Type Malfunction
Event Description
It was reported that the site's charging cradle was broken. Troubleshooting was performed where the ac adapter cable was moved from the back of the cradle, and plugged into the flying lead on the serial cable. This configuration was plugged directly into the hand held rather than using the charging cradle. It was observed that touching or moving the connection caused intermittent charging. The site was sent a new cable to replace the non-working cable. The non-working cable has been retrieved, is en route to mfr for analysis.

Manufacturer Narrative
Device failure is suspected, but did not cause or contributed to death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1706682
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #7 on: January 28, 2017, 03:30:34 AM »

Model Number MODEL 250
Event Date 01/01/2010
Event Type Malfunction
Event Description
It was reported that the physician's hand held was charging intermittently and it is suspected that the ac adapter cord is not working properly. The physician requested a new hand held to replace the non-working device. Good faith attempts to obtain the non-working device. Good faith attempts to obtain the non-working device for analysis have been made, but the device has not been returned to date.

Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1621078
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #8 on: January 28, 2017, 03:31:35 AM »

Model Number MODEL 250
Event Date 04/01/2010
Event Type Malfunction
Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to a death or serious injury.

Event Description
Reporter indicated a vns (b)(6) handheld computer was not holding a charge due to the a/c adapter having a loose connection. The computer has been requested for return, but has not been received to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1720392
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #9 on: January 28, 2017, 03:32:35 AM »

Model Number MODEL 250
Event Date 06/15/2010
Event Type Malfunction
Event Description
Initial reporter indicated that their (b) (6) handheld computer is not turning on. It was reported that it had been plugged in overnight and the power light was green while plugged in. When the handheld was unplugged, the light turned orange. It was plugged back in and the light stayed orange. The handheld was returned for product analysis without the ac adapter and serial cable. Testing on the handheld showed it met specifications. At this time, the issue is believed to be with the ac adapter. Good faith attempts thus far have been unsuccessful in attaining that for product analysis.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1760912
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #10 on: January 28, 2017, 12:00:27 PM »

Model Number 250
Device Problems Break; Failure to charge
Event Date 10/06/2009
Event Type Malfunction
Event Description
It was reported that the sites handheld was broken and that the connector piece in the handheld where the ac adapter is inserted had broken. The handheld would subsequently not charge as a result of the broken connector piece. The handheld, ac adapter, serial cable, db9 adapter, and (b) (4) software flashcard were returned to mfr for analysis. Analysis identified that the connector of the hand held device was damaged and the ac adapter could no longer charge the main battery. The cause of the damage ac adapter is unk, however, it is most likely associated with mishandling of the device. No other anomalies on the device performance were noted during testing using a known good ac adapter or the main battery with a full charge.

Manufacturer Narrative
Device failure occurred, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1542373
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #11 on: January 29, 2017, 02:55:08 AM »

Model Number MODEL 250
Event Date 08/20/2009
Event Type Malfunction
Manufacturer Narrative
Device malfunction is suspected, but did not cause or contribute to a death or serious injury.

Event Description
Reporter indicated a vns therapy programming handheld ac adapter cord was not working. The handheld had been plugged in for several hours and would not charge or turn on. When the handheld was plugged in with a separate power cord, the handheld power light illuminated, charged and turned on as intended. Good faith attempts to obtain device have been unsuccessful to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1502863
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #12 on: January 30, 2017, 01:59:54 AM »

Model Number 250
Event Date 05/04/2009
Event Type Malfunction
Event Description
It was reported to manufacturer that the sites hp hand held ac adapter cord had broken and as a result, they were no longer able to charge the hand held. The site requested a new hand held be sent to replace that non-working device. It is unk how the ac adapter cord broke, but it was reported that "a black piece on the cord broke and fell off". Good faith attempts to obtain the non-working device for analysis have been made, but have been unsuccessful to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1436502
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #13 on: January 30, 2017, 02:00:55 AM »

Model Number 250
Event Date 01/01/2009
Event Type Malfunction
Event Description
Rptr indicated they were having issues with their handheld computer. The device was sent to the manufacturer for analysis. Before analysis was completed info was found that the issue had previously been resolved with troubleshooting. Product analysis was later completed and found that the flashcard was missing a specific file that is meant to be installed on the handheld computer. Product analysis also found that the ac power adapter used to charge the handheld computer was damaged and unable to charge the device. When a known good ac power adapter was used, no other anomalies were identified with the handheld computer.

Manufacturer Narrative
Device failure occurred but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1375769
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #14 on: January 30, 2017, 08:21:13 AM »

Model Number MODEL 250
Event Date 10/01/2008
Event Type Malfunction
Event Description
It was reported that a handheld computer's power adapter cord was physically broken. It will only charge the handheld when held a certain way. A replacement handheld computer was sent to the physician. Good faith attempts to obtain the old handheld for analysis have been unsuccessful to date.

Manufacturer Narrative
Device malfunction is suspected, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1237727
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #15 on: January 31, 2017, 01:54:12 AM »

Model Number 250
Event Date 08/01/2008
Event Type Malfunction
Event Description
Reporter indicated that the gold prongs on the ac adapter cord for his hp jornada handheld computer were broken. Product has been requested, but has not been returned to manufacture to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1172597
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #16 on: January 31, 2017, 01:54:56 AM »

Model Number 250
Event Date 07/14/2008
Event Type Malfunction
Event Description
Initial reporter indicated that they were at a physicians office testing out their new programming system and "she couldn't get the dell x50 handheld computer turn on event when plugged into the wall". When plugged into the wall, the power light did not light up and the lock button was not in the red position. A hard reset and soft reset was performed and the handheld did not turn on. The handheld was returned for analysis with 7. 1 programming software. Both products went through product analysis and no anomalies were identified against the handheld or software performance. The handheld was returned without the ac adapter therefore a malfunction is suspected against the ac adapter. Good faith attempts have been made for product return.

Manufacturer Narrative
Device malfunction suspected, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1148007
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #17 on: January 31, 2017, 01:56:25 AM »

Model Number 250
Event Date 06/01/2008
Event Type Malfunction
Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to death or serious injury.

Event Description
It was reported to mfr that their ac adapter cable for the hand held was frayed and requested that a new one be sent to the site. The non-working cable has been sent back to mfr and analysis is underway.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1081558
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #18 on: February 01, 2017, 07:50:13 AM »

Model Number 250
Event Date 04/30/2008
Event Type Malfunction
Manufacturer Narrative
Device malfunction suspected, but did not cause or contribute to a death or serious injury.

Event Description
Initial reporter indicated that a hospital needed a new power cord for their hp jonada handheld computer, because it was not "holding its charge". Reported "the gold prongs were facing up, and they tried using different outlets. The physician thinks the problem is the ac adapter cord because it is loose. " good faith attempts have been made for product return.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1053320
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #19 on: February 01, 2017, 07:51:09 AM »

Model Number 250
Event Date 03/01/2008
Event Type Malfunction
Event Description
It was reported to mfr that a black piece of plastic that covered the prongs on the ac adapter cable had broken off, which made the hand held incapable of being plugged in to charge. A new cable was sent to the site and the hand held is now able to be charged with no add'l problems. Attempts to obtain the damaged cable have been made, but have been unsuccessful to date.

Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1031298
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #20 on: February 03, 2017, 02:22:21 AM »

Model Number MODEL 250
Event Date 10/01/2007
Event Type Malfunction
Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

Event Description
Reporter indicated the ac adapter that connects to the handheld programming system is sometimes loose due to normal wear and tear. The product was returned to the mfr and a product analysis was performed. Visual inspection identified a broken solder joint between the pcb and a screw mount. The screw mount connects the back of the handheld to the pcb near the main battery connector allowing good electrical contact between the main battery and the handheld device. This anomaly can cause the intermittent issues associated with loss of power while operating the device on battery power. It can also prevent the main battery from being charged using the ac adapter.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=976450
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #21 on: February 03, 2017, 02:23:12 AM »

Model Number MODEL 250
Event Date 09/01/2007
Event Type Malfunction
Manufacturer Narrative
Device failure occurred, but did not cause or contribute to a death or serious injury.

Event Description
Reporter indicated that an hp handheld computer's ac adapter was broken. The connection which inserts into the handheld computer was reported to have been broken completely off. The ac adapter was reported to have been stuck behind a cabinet and could not be returned for analysis.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=922854
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #22 on: February 04, 2017, 03:25:41 AM »

Model Number MODEL 250
Event Date 06/01/2005
Event Type Malfunction
Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

Event Description
Reporter indicated that his vns therapy programming handheld was not operating properly. He stated that "the power cord is broken and the handheld has horizontal lines going across it. " troubleshooting did not resolve the issue. The handheld was subsequently returned to manufacturer for product analysis. The reported complaint associated with the power cord was verified during visual analysis. External visual inspection found that the plug on the ac adapter that attaches to the handheld was broken, and consequently the ac adapter was no longer able to plug into the handheld to charge the main battery. The reported complaint associated with horizontal lines on the screen was also verified. The cause for the lines is associated with loss of power to the handheld device during operation. The horizontal lines on the screen mostly likely appeared as a result of the screen loosing power as the device shut down abruptly. This is expected behavior when the main battery is depleted while the device is in operation.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=900976
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #23 on: February 04, 2017, 03:26:30 AM »

Model Number MODEL 250
Event Date 09/01/2006
Event Type Malfunction
Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

Event Description
Reporter indicated that his vns therapy programming handheld was no longer properly holding a charge. Handheld was subsequently returned to manufacturer for product analysis. An analysis was performed on the handheld and the cause for the complaint was found to be associated with a broken ac adapter. Visual inspection identified that the connector to the hand held device was damaged, and the ac adapter could no longer charge the main battery. No other anomalies on the device performance were noted during testing using a known good ac adapter or the main battery with a full charge.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=902457
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #24 on: February 04, 2017, 03:27:23 AM »

Model Number MODEL 250
Event Date 05/01/2006
Event Type Malfunction
Event Description
Reporter indicated that his vns therapy programming handheld power cord was broken. Handheld and power cord were subsequently returned to manufacturer for product analysis. An analysis was performed on both products and the broken power cord was verified during the visual analysis. It was found that the ac adapter connector to the handheld was broken, and no longer able to provide power to the handheld. No other anomalies with device performance were noted during testing using a known good ac adapter or the main battery with a full charge.

Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=902196
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #25 on: February 04, 2017, 03:28:11 AM »

Model Number MODEL 250
Event Type Malfunction
Event Description
Reporter indicated that her vns therapy programming handheld was not operating properly, in that "it was continually locking up and resetting when it was being used. " it was additionally reported that the handheld power cord was broken. Handheld and power cord were subsequently returned to manufacturer for product analysis. An analysis was performed on the handheld and the cause for the first complaint was found to be associated with a broken solder connection near the main battery terminal. The broken solder connection prevented the main battery from making good electrical contact with the pcb. This condition caused the handheld to intermittently lose power and shut down while operating. During the analysis, visual inspection also verified the complaint that the power cord was broken off, and part of it was still in the connector port. As a result the ac adapter could not longer charge the main battery. The damaged power cord is most likely associated with mishandling of the handheld device.

Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=902407
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #26 on: February 04, 2017, 03:28:57 AM »

Model Number MODEL 250
Event Date 01/01/2007
Event Type Malfunction
Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

Event Description
Reporter indicated that his vns therapy programming handheld power cord was broken. Cord was subsequently returned to manufacturer for analysis. An analysis was performed on the ac adapter and the reported complaint was verified. Visual inspection identified that the connector to the handheld device was damaged, and the ac adapter could no longer charge the handheld main battery.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=902424
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #27 on: February 04, 2017, 03:29:45 AM »

Model Number MODEL 250
Event Date 11/01/2006
Event Type Malfunction
Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

Event Description
Reporter indicated that her vns therapy programming handheld would not properly power on, even after charging it overnight. Handheld was subsequently returned to mfr for product analysis. An analysis was performed on the handheld and the cause for the complaint was found to be associated with a broken ac adapter. Visual inspection identified that the connector to the handheld device was damaged, and the ac adapter could not longer charge the main battery. No other anomalies on the device performance were noted during testing using a known good ac adapter or the main battery with a full charge.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=902436
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #28 on: February 04, 2017, 03:30:33 AM »

Model Number MODEL 250
Event Date 05/01/2006
Event Type Malfunction
Event Description
Reporter indicated that her vns therapy programming handheld was no longer charging correctly. The handheld and accompanying charger were subsequently returned to manufacturer for product analysis. An analysis was performed on the products and the cause for the battery performance issue was found to be associated with a broken ac adapter. Visual inspection identified that the connector to the handheld device was damaged, and the ac adapter could no longer charge the main battery. No other anomalies on the device performance were noted during testing using a known good ac adapter or the main battery with a full charge.

Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=902195
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #29 on: February 04, 2017, 03:31:20 AM »

Model Number MODEL 250
Event Date 07/01/2007
Event Type Malfunction
Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to a death or serious injury.

Event Description
It was reported that a site's handheld ac adapter shorted out when it was plugged into an electrical outlet. A replacement ac adapter was sent. Attempts to have original ac adapter returned have been unsuccessful.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=897927
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #30 on: February 04, 2017, 03:32:10 AM »

Model Number 250
Event Date 07/12/2007
Event Type Malfunction
Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

Event Description
Reporter indicated that her programming system was not operating properly, as it would not communicate with patients' vns therapy pulse generators. Programming handheld was subsequently returned to manufacturer for product analysis. An analysis was performed on the handheld and it was identified that the ac adapter was damaged and no longer able to provide consistent power to the handheld. Once a known good ac adapter was used, no anomalies with the device performance were noted during testing using ac adapter or the main battery with a full charge.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=891275
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #31 on: February 05, 2017, 01:47:46 AM »

Model Number MODEL 250
Event Date 06/01/2007
Event Type Malfunction
Manufacturer Narrative
Device failure occurred, but did not cause or contribute to a death or serious injury.

Event Description
Rptr indicated that a hp handheld computer had a loose ac adapter connection and the handheld would not power on.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=881177
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #32 on: February 05, 2017, 01:48:34 AM »

Model Number MODEL 250
Event Date 01/01/2007
Event Type Malfunction
Event Description
Reporter indicated that the site's hp handheld computer battery was not holding a charge. It was also reported that the ac adapter power cord appeared to be damaged. A new handheld was sent to the site.

Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=871763
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #33 on: February 05, 2017, 01:49:38 AM »

Model Number MODEL 250
Event Date 05/23/2007
Event Type Malfunction
Manufacturer Narrative
Device malfunction is suspected, but did not cause or contribute to a death or serious injury.

Event Description
It was reported that the handheld battery was dead and would not charge when it was plugged into the ac adapter. The ac adapter was making a loose connection with the handheld and one of the gold prongs on the ac adapter had fallen off. A replacement was sent to the site. The handheld was returned to manufacturer. Product analysis is pending.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=868596
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #34 on: February 05, 2017, 01:51:23 AM »

Model Number MODEL 250
Event Date 04/19/2007
Event Type Malfunction
Event Description
Reporter indicated that hp jornada handheld computer had a bad connection where the ac adapter connects to it. It is not making a proper connection and the power button did not come on. Product analysis on the handheld revealed that the support tab of the terminals in the ac adapter connector was broken and that one of the terminals was bent, which would prevent a proper connection. These conditions were most likely the result of the force exerted by user handling.

Manufacturer Narrative
Device malfunction occurred, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=852320
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #35 on: August 02, 2017, 12:11:26 AM »

Model Number 250
Event Date 06/29/2011
Event Type  Malfunction   
Manufacturer Narrative
Device failure occurred, but did not cause or contribute to a death or serious injury.
 
Event Description
It was initially reported that the physician was having issues communicating with a pt's generator and was receiving the error message "there is an error establishing communication with the device, try repositioning the wand". They repositioned the wand, replaced the 9v battery and checked all connections but was still unsuccessful with communicating with the generator. The pt was taken to a different room and was able to be interrogated and it was confirmed that the pt was not at end of service. However, the error message was received again when diagnostics were attempted. The office was eventually able to perform a final interrogation to confirm the pt was at intended settings. Additional info came in indicating that the handheld was not charging. The indicator light would be red if she held the cord but would go off when the cord was released. A new handheld was sent to the site and they were able to use it successfully with their current wand. The handheld and flashcard were returned to the mfr for eval. No anomalies associated with the handheld serial cable were noted during testing. During the analysis, it was identified that the ac adaptor was defective and unable to supply power to the handheld. No further anomalies were identified using a known good ac adapter. No anomalies associated with flashcard software or databases were identified during the flashcard analysis. The flashcard and software performed according to functional specifications.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2199164
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #36 on: August 10, 2017, 07:12:29 AM »

Model Number MODEL 250
Event Date 08/31/2011
Event Type  Malfunction   
Manufacturer Narrative
(b)(4).
 
Manufacturer Narrative
Device malfunction is suspected, but did not cause or contribute to a death or serious injury.
 
Event Description
It was reported that the physician's handheld would not hold a charge, even when left plugged in overnight. The handheld would power on while plugged in; however once it was removed from the wall, the handheld would not come on. The handheld is normally stored at room temperature in the physician's office. The device has not yet been returned to the manufacturer for product analysis.
 
Event Description
The handheld and flashcard were returned to the manufacturer for analysis and analysis has been completed. An analysis was performed on the returned flashcard. No anomalies associated with flashcard software or databases were identified during the analysis. The flashcard and software performed according to functional specifications. An analysis was performed on the returned handheld. No anomalies associated with the handheld performance were noted during testing using a known good a/c adapter or the main battery with a full charge. The handheld performed according to functional specifications. It was noted during product analysis that the a/c adapter for the handheld device was not returned with the handheld; therefore proper function of the a/c adapter cannot be confirmed and a device issue cannot be completely ruled out. Attempts for the return of the a/c adapter have been unsuccessful to date.
 
Manufacturer Narrative
While proper function of the handheld device was confirmed, a device charging issue cannot be ruled out as the a/c adapter for the handheld was not returned.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2263580
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #37 on: August 29, 2017, 12:43:56 AM »

Model Number MODEL 250
Event Date 11/01/2011
Event Type  Malfunction   
Manufacturer Narrative

Event Description
It was initially reported that the physician's handheld would not turn on. The handled was reported to be fully charged and plugged in, but would not turn on. The physician declined to perform troubleshooting. The wand, handheld and flashcard were returned to the manufacturer for evaluation. Product analysis is planned, but has not been completed.
 
Event Description
Additional information was received that product analysis was complete on the wand, handheld and flashcard. No visual or mechanical anomaly was identified with the wand. Continuity testing of the serial data cable and the battery cable passed. The programming wand performed according to functional specifications. The handheld was able to power on and off during testing with no observed anomalies. During the analysis it was identified that the ac adapter was no longer able to provide power to the handheld. A root cause for the ac adapter failure could not be identified during the analysis. No anomalies associated with flashcard software or databases were identified during the flashcard analysis. The flashcard and software performed according to functional specifications.
 
Manufacturer Narrative
Device failure was found, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2384812
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #38 on: August 29, 2017, 12:44:55 AM »

Model Number MODEL 250
Event Date 11/16/2011
Event Type  Malfunction   
Manufacturer Narrative

Event Description
Reporter indicated that a charger power cord for a vns (b)(4) x50 computer was not charging the computer. The power cord ac adapter was returned for analysis. An analysis was performed on the returned ac adapter and the reported allegation was verified. The cause for the reported complaint is associated with a defective 120v ac power cord. Continuity testing of the cord identified that both wire connections were open. A root cause for the open wire connections could not be identified during the analysis. No further anomalies associated with the ac adapter were identified using a known good 120v power cord.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2377525
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #39 on: September 06, 2017, 08:14:06 AM »

Model Number MODEL 250
Event Date 01/03/2012
Event Type  Malfunction   
Event Description
Analysis was completed of the returned handheld. Analysis of the handheld revealed that during the analysis it was identified that handheld was unable to charge the main battery. The primary cause for the anomaly is associated with broken solder connections on the handheld main board. Additionally, it was identified that the returned ac adapter and cradle were unable to power and charge the returned handheld. The cause for the anomaly is associated with an incompatible aftermarket ac adapter. No further anomalies were identified. Moreover analysis of the flashcard indicated that the flashcard and software performed according to functional specifications.
 
Manufacturer Narrative
Device failure occurred, but did not cause or contribute to a death or serious injury.
 
Event Description
It was initially reported by a neurologist that his handheld computer could not be recharged due to unknown reason. There was no user mishandling reported to have contributed to the event. A new handheld was provided to the physician and the reported handheld was returned to the manufacturer for analysis.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2435359
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #40 on: September 12, 2017, 12:02:59 AM »

Model Number MODEL 250
Event Date 03/16/2012
Event Type  Malfunction   
Event Description
It was initially reported by a nurse that, when she plugged the handheld into the wall outlet on (b)(6) 2012, the handheld did not appear to be charging and was at 50% battery life. She then exchanged the ac power cord with the cord from a different handheld and plugged it in to the same wall outlet. The handheld started charging correctly; becoming fully charged after about 20 minutes. A replacement power cord was requested and the old ac adapter was returned for product analysis. Analysis was completed on (b)(4) 2012. An analysis was performed on the returned ac adapter. During the analysis the ac adapter was utilized in an attempt to power a known good handheld device; however this was unsuccessful. The power button did not indicate that the handheld was receiving power from the ac adapter. The cause for the reported complaint is associated with a defective power supply module; however the cause of the defect could not be determined during the analysis.
 
Manufacturer Narrative

Event Description
Further analysis of the power supply, completed on (b)(6) 2012, identified that the most likely cause for the power supply failure is associated with a component failure caused by the contamination of an unknown green substance that was identified inside the case, on the pcb, and on the power supply components.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2529946

Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #41 on: January 10, 2018, 01:23:29 AM »

Model Number 250
Event Date 01/01/2009
Event Type Malfunction
Event Description
Neurosurgeon reported that the adapter cord connecting the dell axim handheld computer and the programming wand was broken. A new handheld was sent to the neurosurgeon and the previously used handheld was returned to the mfr. At this time, the handheld and its components are undergoing product analysis.

Manufacturer Narrative
Device failure is suspected, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1455830
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #42 on: March 12, 2018, 12:20:58 AM »

Model Number MODEL 250
Event Date 06/20/2012
Event Type Malfunction
Manufacturer Narrative

Event Description
An analysis was performed on the returned handheld. During the analysis it was identified that the ac adapter was defective and could no longer power the handheld. No further anomalies associated with the handheld performance were noted during testing using a known good ac adapter or the main battery with a full charge. When analysis of the software was performed it was identified that the flashcard contained a non-vns file and non-vns folders. The file and folders were created as a result of the flashcard being inserted into a (b)(6) computer. The file and folders had no impact on the vns and flashcard performance. No anomalies associated with flashcard software or databases were identified during the flashcard analysis. The flashcard and software performed according to functional specifications.

Manufacturer Narrative
Device malfunction occurred but did not cause or contribute to a death or serious injury.

Event Description
It was reported by a company representative that a physician's handheld computer no longer worked. Additionally, the area representative indicated he used his programming wand with the reported handheld and communication was unsuccessful with a vns patient. The reported handheld was returned to the manufacturer and at the moment remains under analysis.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2653531
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #43 on: March 18, 2018, 12:27:03 AM »

Model Number MODEL 250
Event Date 10/03/2012
Event Type Malfunction
Event Description
Additional analysis was performed on the returned ac adapter. The returned ac adapter was opened to perform a visual inspection of the internal components and printed circuit board. Results of the visual inspection identified no anomalies. No further analysis could be performed as this is a non-cyberonics manufactured product and no additional product information is available.

Event Description
Reporter indicated a vns dell x50 computer would not power on. The computer used to power on but would take 10-15 minutes to do so. Now, it does not power up at all. A hard reset of the computer did not resolve the issue, and it was confirmed the "lock" button was not engaged. The computer, flashcard, and programming wand were returned for analysis. No anomalies were noted with the programming wand or flashcard. No anomalies associated with the main battery of the computer were identified during the analysis. During the analysis it was identified that the returned ac adapter was defective and unable to power the returned handheld. The cause for the anomaly is associated with an incompatible (underrated) ac adapter. No further anomalies were identified.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2804113
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #44 on: April 28, 2018, 12:21:44 AM »

Model Number MODEL 250
Event Date 05/21/2013
Event Type  Malfunction   
Event Description
On (b)(6) 2013 it was reported that the handheld was not charging properly. It was stated that the power cord was found to be the problem as the handheld worked fine with an alternative charger. It was stated that the wand and cable adapter are all fine. The nurse said she had dropped the power cord and it hit the floor at the ¿part of the box which connects both the wires¿; since that time it has not worked. When it is connected to the handheld and plugged in there is no charging light at all. It was stated that the device would be returned to the manufacturer for product analysis but it has not been received to date.
 
Manufacturer Narrative

Event Description
It was reported that the device would be returned for product analysis but it appears to have been lost in the mail as it was reported to have been sent but it was not received by the manufacturer for product analysis.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3167335
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #45 on: July 25, 2018, 05:48:02 AM »

Model Number 250
Event Date 11/01/2010
Event Type  Malfunction   
Manufacturer Narrative
Device failure occurred, but did not cause or contribute to a death or serious injury.
 
Event Description
It was reported that the power cable on a physician's hp jornada handheld device was not making a good connection with the handheld; therefore, the handheld could not charge fully. A replacement handheld and power cable was sent to the physician and the hp jornada and power cable in question were returned to the manufacturer for analysis. During the analysis oft he handheld, it was identified that the connector to the handheld device was damaged and the ac adapter could no longer charge the main battery. No other anomalies on the device performance were noted during testing using a known good ac adapter or the main battery with a full charge. No anomalies associated with flashcard software or databases were identified during the flashcard analysis. The flashcard and software performed according to functional specifications.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=1931687
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #46 on: October 13, 2018, 04:25:38 AM »

Model Number MODEL 250
Event Date 06/15/2012
Event Type  Malfunction   
Event Description
A vns consultant was at a clinic, and it was reported that their handheld computer was not holding a charge. The consultant could not get the handheld computer to turn on even with it plugged in. It was verified that the lock button was not on, and that a hard reset did not bring up the screen. The programming system is kept in the normal office environment temperature. The handheld was returned, and analysis is pending completion.
 
Manufacturer Narrative

Event Description
The handheld computer was received with a v8. 0 flashcard and with the main battery depleted. The handheld was received without an ac adapter and serial cable. The handheld device was powered using a known good ac power supply with no anomalies from inhouse. The handheld's main battery was fully recharged successfully using the power supply adapter from inhouse. The power button light on the handheld computer was amber and then later turned green giving the indication that the main battery was fully charged. The back-up battery indicator read "normal". This is an indication that the backup battery also was recharged successfully. No anomalies associated with the handheld performance were noted during testing using the ac adapter or the main battery with a full charge. The handheld performed according to functional specifications. It is likely that their was an issue with the ac adapter cord charging the handheld at the office. The cord was discarded, therefore will not be returned for analysis.
 
Manufacturer Narrative
Event likely related to discarded ac adapter cord failure. Device malfunction suspected with ac adapter cord that was not returned and discarded.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2640276
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #47 on: October 19, 2018, 11:12:35 AM »

Model Number MODEL 250
Event Date 08/02/2012
Event Type  Malfunction   
Event Description
Product analysis was completed for both the handheld computer and related software. The ac adapter was not returned for analysis. During the analysis it was identified that handheld was unable to charge the main battery. The cause for the anomaly is associated with broken solder connections on the handheld main board. Also during the analysis it was identified that the main battery was inserted backwards. This anomaly can also prevent the handheld from charging the main battery. No further anomalies were identified. No anomalies associated with flashcard software or databases were identified during the flashcard analysis. The flashcard and software performed according to functional specifications.
 
Event Description
It was initially reported on (b)(6) 2012 that the physician's ac power cord for the dell x50 handheld computer was damaged due to somebody inadvertently slamming the cord in a drawer. It became frayed and exposed, so they requested a replacement ac power cord/adapter. The company representative confirmed that the damage was due to the site's handling and they were not alleging a device malfunction. A replacement was provided, but later on (b)(6) 2012, it was reported that the computer would not charge. It was initially thought it was a power cord issue, however even with the power cord, the handheld would not charge and thus the battery was depleted. A replacement handheld computer was provided to the site, and the troubled computer was returned to the manufacturer for product analysis. No mishandling was suspected, and the programming system is stored at room temperature in the provided nylon sack plugged into the ac power outlet. Product analysis has not been completed to date on the returned computer and related software. It was noted upon receiving that the battery was inserted backwards into the computer.
 
Manufacturer Narrative

Manufacturer Narrative
Device failure occurred, but did not cause or contribute to a death or serious injury.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=2746945
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #48 on: March 18, 2019, 01:31:16 AM »

Model Number MODEL 250
Device Problems Failure to Interrogate ; Failure to Power Up
Event Date 08/31/2015
Event Type  Malfunction   
Event Description
On (b)(6) 2015, it was reported that the physician's handheld will not turn on despite being plugged into a known to be working outlet. The problem was noted to have begun two weeks prior and multiple patients were unable to be interrogated. The handheld and flashcard were returned for product analysis on (b)(4) 2015. Product analysis is still underway and has not yet been completed.
 
Event Description
On (b)(6) 2015 product analysis was completed on the handheld and flashcard. No anomalies associated with the main battery were identified during the analysis. During the analysis it was identified that the serial cable was providing power to the handheld intermittently. The cause for the anomaly is associated with an open electrical connection in the power cable portion. As a result of the open wire connection, the handheld would receive power from the ac adapter intermittently. No other anomalies were identified. No anomalies associated with flashcard software were identified during the flashcard analysis. The flashcard and software performed according to functional specifications.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=5137698
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #49 on: March 21, 2019, 12:26:15 AM »

Model Number MODEL 250
Device Problem Premature Discharge of Battery
Event Date 10/09/2015
Event Type  Malfunction   
Event Description
It was reported that the handheld's battery would not hold a charge. The handheld has been received for analysis, which is underway but has not been completed to-date. No additional information has been received to-date.
 
Event Description
Analysis was performed on the returned handheld. No anomalies associated with the main battery were identified during the analysis. During the analysis, it was identified that the ac adapter was damaged and unable to power the handheld. The cause of the damage was associated with mishandling of the device. No further anomalies were identified. An analysis was performed on the returned flashcard and no anomalies associated with the flashcard software or databases were identified during the analysis. The flashcard and software performed according to specifications.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=5202415
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #50 on: April 03, 2019, 12:48:54 AM »

Model Number MODEL 250
Event Date 04/11/2016
Event Type  Malfunction   
Manufacturer Narrative
 
Event Description
It was reported that the physician's handheld device will not power on. This occurs whether or not the device's charging cable is plugged in. No additional pertinent information has been received to date.
 
Event Description
The suspect handheld device and associated flashcard were received by the manufacturer on 05/05/2016, and product analysis was subsequently completed. For the returned handheld, visual inspection of the device showed that the screen was cracked. Additionally, the lock button was in the locked position, thus the initial attempt to power on the handheld was unsuccessful. The lock button was moved to the unlocked position, and the device was successfully powered on. An attempt to advance past the welcome screen was performed but was unsuccessful, as the touchscreen was unresponsive. The cracked screen was replaced with a known good screen, and the initial setup process was performed. No further issues were noted with device charging, generator interrogations or diagnostics. The cause of the cracked screen was mechanical stress. The analysis performed on the returned flashcard showed no identified issues. No anomalies associated with flashcard software or databases were identified during the flashcard analysis. The flashcard and software performed according to functional specifications.
 
Manufacturer Narrative
Describe event or problem, corrected data: follow-up report #1 inadvertently omitted section of product analysis results that pertained to portions of device not returned.
 
Event Description
Further review of product analysis results revealed that the ac adapter cable was not returned with the handheld. As product analysis was not completed on this portion of the device, the possibility the failure to power on was from an issue with the power cord could not be eliminated. Follow up with the physician's office indicated that the power cable could not be located and returned to the manufacturer for product analysis. It has not been returned to date. No additional pertinent information has been received to date.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=5626315
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #51 on: April 03, 2019, 12:49:34 AM »

Model Number MODEL 250
Event Date 04/11/2016
Event Type  Malfunction   
Event Description
It was reported that company representative's tablet is faulty. The tablet didn't turn on despite good charging of the device battery, other tablet with was connected with his own wand and usb white cable and the programming system worked fine. It was also reported that no switch on of the tablet is possible also when it is plugged into the wall outlet. The review of manufacturing records confirmed that the tablet passed all functional tests prior to distribution. The return of the suspect tablet is expected but it has not been received to date.
 
Manufacturer Narrative
Date when the manufacturer's employee was aware about the event; corrected data: the previously submitted mdr inadvertently provided an incorrect initial aware date.
 
Event Description
The suspected tablet was returned to the manufacturer on 05/19/2016. An analysis was performed on the returned tablet and the reported allegation was not verified. No anomalies associated with the tablet performance were noted during testing using the ac adapter or the main battery with a full charge. The tablet performed according to functional specifications.
 
Event Description
It was noted during the analysis of the returned tablet that the main battery depleted; therefore the ac power adapter was suspected to be faulty. Attempts to the user, to return the ac power adapter, were unsuccessful. It was reported that suspected adapter was discarded.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=5643645
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62836


« Reply #52 on: April 11, 2019, 01:57:18 AM »

Model Number MODEL 250
Event Date 06/22/2016
Event Type  Malfunction   
Manufacturer Narrative
(b)(4).
 
Event Description
It was reported on (b)(6) 2016 that a physician's tablet was not charging with the charger. The company representative used her charger with the physician's tablet, and the tablet charged properly. The company representative was the one to identify that the tablet was not charging. She also tried using a different outlet with the charger, but it did not resolve the issue. The tablet has not been received to date.
 
Event Description
The ac adapter was received on 07/26/2016. Analysis was approved on 08/18/2016. Analysis of the returned ac adapter verified that it was no longer able to power a known good tablet. The cause for the device failure could not be determined during the analysis.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=5801596
Logged