« Reply #60 on: April 10, 2014, 08:09:18 AM »

Model Number 37601
Event Type  Injury   Patient Outcome  Other
Manufacturer Narrative
Concomitant medical products: product id: 37085-40, serial# (b)(4), implanted: (b)(6) 2010, product type: extension. Product id: 37085-40, serial# (b)(4), implanted: (b)(6) 2010, product type: extension. Product id: 3387s-40, lot# v499735, implanted: (b)(6) 2010, product type: lead. Product id: 3387s-40, lot# v460686, implanted: (b)(6) 2010, product type: lead. (b)(4).

 
Event Description
It was reported that the patient had 10 ultra sound treatments beforehand and 2 weeks ago had 6 procedures. It was noted that the patient had a loss of therapeutic effect. It was noted that the reporter stated that with all the patient¿s medications the patient lost their balance all the time and fell frequently. It was noted that when the patient fell they did not hit their head and fell on their elbow and broke it. It was noted that the patient also had experienced changes in mental cognitive abilities. It was noted that again it was worse. It was noted that the patient had uncontrollable movements, and made spontaneous drastic movements, with their hands and sometimes hit the wall for furniture. It was noted that the patient had gone to the health care professional (hcp) and was referred for physical therapy. It was noted that the hcp was checking the device. Additional information received reported that the patient recently had an ultrasound and was not feeling well. It was noted that the ultrasound was five days ago and the patient had a convulsion like epilepsy and was shaking. It was noted that the patient had an ultrasound because of the parkinson¿s. It was noted that the patient had hit their elbow and the elbow had a bump so the patient was referred for ultrasound. It was noted that it was therapeutic ultra sound. It was noted that patient had had this before when the weather was changing and was thinking it was maybe the ultrasound procedure. Additional information has been requested, but it was not available as of the date of this report.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3667093
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #61 on: April 10, 2014, 08:10:17 AM »

Model Number 37601
Event Type  Injury   Patient Outcome  Required Intervention
Event Description
It was reported that the patient had more tremors in the last 2 weeks and had a tendency to fall, sometimes 2-3 times a day with some of them noted as ¿quite heavy. ¿ in addition, it was reported the patient could not connect with the implantable neurostimulator (ins) using the programmer unit. This issue was resolved when the antenna accessory was unplugged from the programmer (although it was noted that the connection also worked with the antenna plugged into the programmer). Follow up information received from the healthcare professional (hcp) on (b)(6) 2013 reported that cause of the event was likely urinary tract infection (uti), severe parkinson¿s disease, and sub optimal placement of the lead components of the ins system. If additional information is received, a follow up report will be sent.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3672833
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #62 on: April 10, 2014, 08:10:54 AM »

Model Number 37612
Event Type  Injury   Patient Outcome  Disability
Event Description
It was reported that while stimulation was turned on the patient¿s equilibrium was off and he had been falling forward and to the left for the past 6 months prior to the date of this report. It was noted that the patient saw the healthcare professional every 6 months and the healthcare professional had mentioned that it was a possible side effects from surgery. It was further noted that the patient had looked up possible equilibrium after deep brain stimulator surgery and it stated it was possible. When the implantable neurostimulator (ins) was off the patient was not feeling his equilibrium off but the shaking started back. It was noted that because of the patient¿s condition he was getting a wheelchair. Additional information was requested but had not been received as of the date of this report.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3686601
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #63 on: May 11, 2014, 10:14:15 PM »

Model Number 37601
Event Type Injury Patient Outcome Required Intervention
Manufacturer Narrative
Concomitant products: product id: 3387s-40, lot# v883770, implanted: (b)(6) 2012, product type: lead. Product id: 3387s-40, lot# v883770, implanted: (b)(6) 2012, product type: lead. Product id: 37085-40, serial# (b)(4), implanted: (b)(6) 2012, product type: extension. Product id: 37085-40, serial# (b)(4), implanted: (b)(6) 2012, product type: extension. (b)(4).
Event Description
It was reported the patient¿s deep brain stimulation (dbs) device was implanted (b)(6) of 2012 and they could never program it. The leads were implanted and ten days later they put the battery pack in and they waited another ten days before they turned it on to program it. The patient¿s healthcare provider (hcp) could not figure out why on the lowest setting the patient could not talk, his left arm would come up to his chest when he would walk and his right shoulder kept twitching or spasming. From the time it was implanted the device did not really function for the patient; every time they turned it on and gave him medications to try to control the dyskinesia it did not work and this was after a year of oral medication. It was noted the patient was dyskinetic the entire time the device was implanted. It stayed in for about a year until they could not figure out why or what was going on. Every time the patient turned it on ¿patient shoulder would, he was falling three times a day. ¿ the patient fell in his house and broke his ribs the last saturday in (b)(6) 2012. The patient¿s dbs device was turned off and he was better but he still had some of the dyskinesia and had tremendous headaches. In (b)(6) of 2013, the patient had his device removed as he could not handle it anymore. The patient was sure there was a short circuit or something like that which caused the dyskinesia. It was reported the patient thought the leads down in the globus thalamus would not do that, he understood it may affect his speech, but not the rest of it. It was reported the device was leading to such headaches and such problems; he had a big bony growth behind his ear in his skull which was occurring during the entire time the device was implanted. After the device was removed those symptoms went away. Additional information received two weeks later reported somewhere around the second week of (b)(6) 2012 the patient was in severe dyskinesia one night and he could not turn it off. He did not know what was going on and he finally got in touch with a manufacturer representative from (b)(4) who walked his wife through turning his device off. It was also reported the patient¿s unit did not come out easily, the leads and all that stuff did not come out easily. Additional information has been requested but was not available as of the date of this report; a follow-up report will be sent if information becomes available.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3745721
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #64 on: July 04, 2014, 10:45:29 PM »

Model Number 37602
Event Type Injury Patient Outcome Required Intervention
Manufacturer Narrative
Product id 7426, serial# (b)(4), implanted: 2007 (b)(6), explanted: 2013 (b)(6); product type implantable neurostimulator product id 7426, serial# (b)(4), implanted: 2007 (b)(6); product type implantable neurostimulator product id 748251, serial# (b)(4), implanted: 2002 (b)(6); product type extension product id 3387-40, lot# j0211551v, implanted: 2002 (b)(6); product type lead product id 3387-40, lot# j0214131v, implanted: 2002 (b)(6); product type lead product id 748251, serial# (b)(4), implanted: 2002 (b)(6); product type extension product id 7438; product type programmer, patient product id 748251, serial# (b)(4), implanted: 2002 (b)(6); product type extension product id 3387-40, lot# j0211551v, implanted: 2002 (b)(6); product type lead product id 748251, serial# (b)(4), implanted: 2002 (b)(6); product type extension product id 748251, serial# (b)(4), implanted: 2002 (b)(6); product type extension product id 3387-40, lot# j0211551v, implanted: 2002 (b)(6); product type lead product id 748251, serial# (b)(4), implanted: 2002 (b)(6); product type extension product id 3387-40, lot# j0214131v, implanted: 2002 (b)(6); product type lead. (b)(4).
Event Description
It was later reported that the patient was not doing well. The patient was ¿doing the same falling. ¿ patient loses her balance, crosses her feet and her legs freeze causing her to fall. It was noted that this had been going on for ten years prior to the date of this report. The patient would fall 2 or 3 times a day. It was noted that as a result of the falls the patient had had to have knee replacements and knee cap surgery. The patient had a walker but if she had to grab or anything she fell. It was noted that when the first implant was put in 2002 it had helped with her walking. The battery was replaced in 2007 and it ¿did good¿ for about a week and then the patient all of a sudden started having trouble walking. In 2013, one of the batteries was replaced but the other had shown no wear and it was unknown how or why. The device did help with the patient¿s tremors. The patient did not know if she wanted the lead fixed because, she had been in the hospital for 2 weeks and spent 3 months in a nursing home recovering. It was noted that the right battery was the one that was replaced in 2013 and the left was the one that had shown no use, left battery was like brand new. Reference manufacturer¿s report number: 3004209178-2012-10670.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3897360
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #65 on: July 04, 2014, 11:56:04 PM »

Model Number 7426
Event Type Injury Patient Outcome Other
Event Description
It was reported the patient was getting ¿zapped¿ in their head at the time of report. It was stated it had been going on for 2 years prior to report. It was noted a few weeks prior the patient started to have seizures more often and they thought the device might have contributed to that. It was stated the patient fell more often and they were losing strength in their legs.
Manufacturer Narrative
Product id: 3387-40, lot# j0104141v, implanted: (b)(6) 2001, product type: lead. Product id: 7495-51, serial# (b)(4), implanted: (b)(6) 2001, product type: extension. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3869665
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #66 on: August 10, 2014, 06:14:25 AM »

Model Number 37601
Event Date 05/15/2014
Event Type Injury Patient Outcome Other
Event Description
It was originally reported on (b)(6) 2014 that the patient fell on the day of the report and hit her back on the edge of a table. The area was really tender and the patient was going to go to urgent care and wanted to check if she could have an x-ray. The patient noted that it was not due to her device or therapy, but due to her parkinson¿s. The patient had a tendency to fall every once in a while without knowing she was going to. About two and a half weeks later it was reported that the patient still had concerns regarding her device or therapy but was working with her doctor or manufacturer representative. The patient listed an appointment for (b)(6) 2014. About three weeks later it was reported that the patient had an increased number of falls since implant. The patient noted that she was falling even before implant, but the falls tended to be more controlled and ¿softer. ¿ the patient stated that before she would ¿just get bruises¿ and was now falling twice as much. The patient thought it may be even in the wake of turning on the implantable neurostimulator (ins) a month after implant. The patient had a fall at easter time. The patient then had a fall on (b)(6) 2014 and two in the ¿last week. ¿ one when the patient was coming in the door, stubbed her foot, and fell and hit her rib. The patient was diagnosed with a cracked rib. The patient then fell again on (b)(6) 2014 and fractured vertebrae l1 by ct analysis at an emergency room (er). At the time of the report the patient fell flat on her behind while standing in the kitchen. The patient was uncertain if she was bending over and noted her dog was there, so she really was unsure how that fall happened. The patient had been on pain medication since the fall and went out and did some yard work the day prior and that seemed to aggravate things. The cause of the event seemed to be temporally related to the surgery. It was a brand new battery, no depletion, and no issues with the lead or extension. It may be worse since turning on the ins and programming, so the patient was asked to turn off the ins and come in for reprogramming on (b)(6) 2014. The patient was also asked to use adaptive equipment to prevent falls, specifically a cane or walker. The patient had an injury and was in recovery, non-life-threatening, but clearly at risk for more falling and there was concern regarding her balance and gait and potential for future falls and injury.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3915421
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #67 on: August 11, 2014, 01:27:59 AM »

Model Number 37602
Event Date 07/09/2014
Event Type Injury Patient Outcome Required Intervention
Event Description
It was reported, the patient had fallen two days prior to report and chipped their tooth and they may lose one of their front teeth because of the fall. It was noted they had been put back on antibiotics to prevent an infection from forming in the patient¿s mouth. They had been off of antibiotics from the possible urine infection for 2-3 weeks before they were put on the antibiotics again for their mouth. The patient had a number of falls ¿lately. ¿ it was noted they had a near fall the day prior to report and a major fall 2 days prior. The patient started to take falls during the summer time. It was noted, the issues with the dbs batteries did contribute to the falls.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3949234
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #68 on: September 06, 2014, 01:48:36 AM »

Model Number 37603
Event Type Injury Patient Outcome Other
Event Description
It was reported that the patient had had balance issues and trouble with speech since implant. It was noted that the patient had fallen a couple of times. Initially it was thought that the patient¿s medication was off but he had been off the medication and was still having problems. Patient had been instructed to turn stimulation down and had further programming appointment scheduled for the friday following the date of this report. Patient¿s stimulation was down from 3. 30 and 4. 10 to 3. 0 and 3. 40. Patient¿s status was unknown. It was later reported that there was disappointment and dissatisfaction expressed about the therapy as far as what had been sold and presented to them prior to implant and what had happened in reality. Patient had tremors prior to the deep brain stimulator implant. Patient had been pretested and it had worked for him. Almost 2 months after implant the patient had a lot of issues. Patient¿s quality of life had not increased. Patient had an appointment scheduled for (b)(6) 2014. Patient was experiencing symptoms. It was noted that the patient had expected it to be an outpatient surgery and instead patient had spent 10 days in rehab. The patient had been told about the risk of infection and stroke but nothing else. The patient had been trained on using the patient programmer but did not know what to do with it. The patient was not very discouraged and not positive. The patient was not enjoying things in life anymore. It was noted that prior to deep brain stimulator surgery the patient had only had some tremors and had not been that advanced with parkinson¿s disease symptoms. The deep brain stimulator surgery was supposed to make the patient better but had made him worse. Patient was not able to return to his physical job. Patient¿s symptoms were change in gait, lethargy, and seizures. It was further noted that the manufacturing representative had been there when the patient had a seizure and the representative had come to rehab. The patient had had a seizure right after surgery at the hospital on the date of implant. At rehab the patient had been completely out of it and had been unresponsive. The healthcare professionals were blaming it on medications. Patient was put on 4 different seizure medications, one being dilantin. Patient no longer took seizure medication. Patient was unresponsive and was sent for a computerized tomography scan (cat scan) which had determined the patient had seizures and that was when patient was put on seizure medications. The healthcare professionals had stated ¿oh no, he should be responsive. ¿ it was noted that nobody seemed to know what was going on. They had come for physical therapy and were not able to wake him up, he had been completely out of it. An electroencephalogram (eeg) was done and had not found anything particular. Following deep brain stimulator surgery patient had the following symptoms: speech issues, gait issues, balance problems, patient had fallen several times, extremely tired, swallowing issues, patient looked at food as if he did not know how to put a fork into it, and patient looked like he had aged (b)(6). Patient used to work full time, go to ymca 3 times a week and drove a car, patient was no longer able to drive or do anything. It was hard to leave the patient alone. It was believed that symptoms were device related because patient had no chemicals in his body except some medication for parkinson¿s and it was not a lot. Healthcare professionals stated they had never seen those symptoms after a deep brain stimulator surgery. Patient was told that this kind of reaction was very common and it might take 6 months to a year to recover from the surgery. Patient had been given a limit on the patient programmer and could not go beyond ¿4 or something. ¿ patient had been adjusting therapy all the time on his own. Patient got better here and there but still had a lot of issues going on and was no longer positive. Patient was scheduled to see the healthcare professional on friday following the date of this report. Additional information received reported the patient was having some gait and speech issues related to programming. Patient had primary tremor with parkinson¿s and as the patient would get tremor control he would get some speech issues. The patient had since been reprogrammed to a bipolar setting and although he had a little bit of a breakthrough tremor his balance was better and so was his speech the last time the manufacturing representative had spoken to the patient. It was unknown if there were any other updates. The manufacturing representative had been involved with the patient through rehab. It was noted that the patient¿s family was anxious about getting the patient¿s progress and getting him programmed so he could go back to work. As of the week prior to the date of this report the patient was doing better and was planning on returning to work. Additional information received reported the alleged seizure was not the reason for rehab and all of their patient¿s go through a rehab cycle. Seizure was not confirmed and had been discussed after the fact. The patient was having some cognitive issues after the battery placement that had led the healthcare professional to believe it may have been a seizure. Patient had been placed on seizure medication and had not had any seizure activity after the first alleged one. Additional information was requested but had not been received as of the date of this report.
Manufacturer Narrative
Concomitant medical products: product id: 37603, serial# (b)(4), implanted: (b)(6) 2014, product type: implantable neurostimulator. Product id: 37642, serial# (b)(4), product type: programmer, patient. Product id: 3708660, serial# (b)(4), implanted: (b)(6) 2014, product type: extension. Product id: 3708695, serial# (b)(4), implanted: (b)(6) 2014, product type: extension. Product id: 3389s-40, lot# va0jcu8, implanted: (b)(6) 2014, product type: lead. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=3979081
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #69 on: September 06, 2014, 05:04:24 AM »

Model Number 37603
Event Type Injury Patient Outcome Other
Manufacturer Narrative
Concomitant medical products: product id: 37642, serial# (b)(4), product type: programmer, patient. Product id: 37603, serial# (b)(4), implanted: (b)(6) 2013, product type: implantable neurostimulator. Product id: 3708660, serial# (b)(4), implanted: (b)(6) 2013, product type: extension. Product id: 3387s-40, lot# va0arp9, implanted: (b)(6) 2013, product type: lead. Product id: 3387s-40, lot# va0bbqp, implanted: (b)(6) 2013, product type: lead. Product id: 3708660, serial# (b)(4), implanted: (b)(6) 2013, product type: extension. (b)(4).
Event Description
It was reported that the patient¿s hands had started to clasp together during the night since (b)(6) 2013. After that, three months after implant, the patient began falling a lot. The patient fell once or twice a day and in (b)(6) 2014 he broke his leg. The patient had not fallen since that time. The patient had lost a lot of coordination and stumbled when he walked. The patient would hold on to something when he walked to keep himself more stable. This issue had been happening for ¿a while,¿ but the patient could not provide an exact date, and he thought his coordination could be getting worse. The healthcare provider (hcp) thought the cause of the event was gait ataxia or possible orthostatic hypotension with polyneuropathy. The therapy was still working for the patient¿s tremors. The patient had an appointment scheduled with his implant surgeon for (b)(6) 2014. Medical intervention and the patient outcome were not reported, so additional information was requested. If additional information is received, a supplemental report will be sent. Refer to manufacturing report #3004209178-2014-15622 as the patient had two ins¿s and it was not specified which was the issue.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4024800
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #70 on: September 06, 2014, 06:20:22 AM »

Model Number 37601
Event Type Injury Patient Outcome Required Intervention
Manufacturer Narrative
Concomitant medical products: product id: neu_unknown_lead, lo# unknown, serial# unknown, implanted: (b)(6) 2010, product type: lead. Product id: 37085-60, serial# (b)(4), implanted: (b)(6) 2010, product type: extension. Product id: 37085-60, serial# (b)(4), implanted: (b)(6) 2010, product type: extension. Product id: neu_ptm_prog, serial# unknown, product type: programmer, patient. (b)(4).
Event Description
It was reported that the patient had needed an implantable neurostimulator (ins) change out the second week of (b)(6) prior to the date of this report. During the week of (b)(6) 2014, the patient had started having therapy issues which had involved more freezing episodes and falling frequently. The patient had fallen once weekly during that time. It was noted that since the ins change out those problems had been less of an issue. The healthcare professional ran impedances on the system and had told the patient that they were normal. The patient had inquired if loss of efficacy was normal toward end of service (eos). Patient¿s therapy was better but not completely back to where it was before the therapy issues had begun. Additional information was requested but had not been received as of the date of this report.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4043778
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #71 on: October 12, 2014, 03:17:48 AM »

Model Number 37601
Event Date 08/06/2014
Event Type Injury
Event Description
It was reported the parkinson¿s disease patient had ¿not been able to have effective stimulation since the implant¿ and that ¿she was worse off with her symptoms¿ following the replacement of her previously depleted implantable neurostimulator (ins). The patient had met with her physician and was told ¿it was the implanted battery that was not working¿ and that ¿they were not able to adjust stimulation. ¿ despite this, it was noted the physician had ¿cranked the stimulation up from 3. 9 to 4. 8 (volts)¿ with ¿no change¿ in the patient¿s symptoms. The patient was ¿unable to get up on her own, she was constantly freezing due to the condition, her head was falling to the side and freezes until she is assisted¿ and she had ¿lost weight¿ and had ¿no appetite¿ at the time of report. The patient ¿could not walk by herself¿ either at that time. It was reported the patient experienced ¿lethargy, loss of reflexes, paralysis, and less than 50% therapy relief¿ that was associated with the event. Information regarding interventions and the patient¿s outcome was not available. Additional information has been requested. A supplemental report will be filed if additional information is received.

Manufacturer Narrative
(b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4118705
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #72 on: October 12, 2014, 06:10:22 AM »

Model Number 37601
Event Type Injury
Event Description
It was reported that the patient had been falling for the past 3 months prior to the date of this report. The patient had taken some pretty hard falls and had been falling about 2-3-5 times a week. The patient had been losing her balance. The patient wanted to know if everything was ok with therapy. The last time the patient was in the healthcare professionals office was in (b)(6) prior to the date of this report and nothing was wrong when checked. There was a patient injury. Following the patient¿s worst fall she had spent 6 weeks in rehab for a cracked pelvic bone. The patient had fell two days prior to the date of this report and hit her head in the bathroom and had gotten a black eye which looked awful. The patient had had more than that and had hit her shoulder. The patient had a speech therapist she was working with and thought she was better that way for speech. The patient had lost the controller since (b)(6) prior to the date of this report. The patient had an appointment with her healthcare professional scheduled in (b)(6). No interventions or outcome were reported regarding this event. Further follow-up is being conducted to obtain this information. If additional information is received a follow-up report will be sent.

Manufacturer Narrative
Concomitant products: product id 37085-60, serial # (b)(4), implanted: (b)(6) 2010, explanted: (b)(6) 2013, product type extension; product id 37642, serial # (b)(4), product type programmer, patient; product id 37085-60, serial # (b)(4), implanted: (b)(6) 2010, product type extension; product id 3389s-40, lot # v436025, implanted: (b)(6) 2010, product type lead; product id 3389s-40, lot # v411494, implanted: (b)(6) 2010, product type lead. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4096758
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #73 on: October 12, 2014, 12:34:57 PM »

Model Number 37601
Event Type Injury
Event Description
It was reported that there was a loss of therapeutic effect. Patient was implanted with the deep brain stimulator about 3 years prior to the date of this report. The patient was quickly going back to shaking like prior to implant. The patient had been falling, dropping things and shaking lightly which had started about 3-4 months prior to the date of this report. It was noted that it had gradually gotten worse and worse and then the patient had started falling. On friday prior to the date of this report, the patient had fallen twice and he could not get up and was unable to get inside which was the worst he had fallen. The patient hurt his back and he had laid there for 3 hours. Since sometime towards to the end of (b)(6), the patient¿s shaking had gotten a little worse and he was unable to hold a glass in his left hand. It was noted that it appeared he would have a new implant sometime in (b)(6), as they were trying to get him in for surgery because ¿this thing was getting worse and worse. ¿ the patient had never turned his stimulation off before. The patient was unable to adjust stimulation after his granddaughter got a hold of his programmer and turned his stimulation down to 0 volts on 1 side, and he was now unable to turn it back up again. When the patient would try and increase stimulation, he would get an information screen but was unable to describe the exact picture on the screen. During the phone call the patient ended up adjusting something and was now able to increase the stimulation and was at 5. 3 volts on the left side and 4. 10 volts on the right, and the issue was resolved. No interventions or outcome were reported regarding this event. Further follow-up is being conducted to obtain this information. If additional information is received a follow-up report will be sent.

Manufacturer Narrative
Product id 37085-60, serial# (b)(4), implanted: 2011 (b)(6); product type extension product id 37085-60, serial# (b)(4), implanted: 2011 (b)(6); product type extension product id 3387s-40, lot# v833497, implanted: 2011 (b)(6); product type lead product id 3387s-40, lot# v779188, implanted: 2011 (b)(6); product type lead product id 37642, serial# (b)(4); product type programmer, patient. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4089335
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #74 on: October 13, 2014, 12:31:23 AM »

Model Number 37603
Event Type Malfunction
Manufacturer Narrative
Concomitant products: product id 37603, serial # (b)(4), implanted: (b)(6) 2012, product type implantable neurostimulator; product id 3387s-40, lot# va01d5k, implanted: (b)(6) 2012, product type lead; product id 37642, serial # (b)(4), product type programmer, patient; product id 3708640, serial # (b)(4), implanted: (b)(6) 2012, product type extension; product id 3387s-40, lot # va01ngu, implanted: (b)(6) 2012, product type lead; product id 37085-40, serial # (b)(4), implanted: (b)(6) 2012, product type extension; product id 37642, serial # (b)(4), product type programmer, patient. (b)(4).

Event Description
It was reported the patient had fallen and broke their shoulder on (b)(6) 2014 and had surgery the following day. They had therapy for about 3 weeks and things were progressing. They thought their symptoms after surgery were worse than before but the last 2 or 3 days prior to report their symptoms had gotten much worse. They had worked with a physical therapist at home on the day of report and could barely walk across the room. They had more shaking getting started from a dead stop. There was a patient injury and a loss of therapeutic effect. It was noted the patient programmer batteries were dead and they had never used the programmer in two years. The patient changed the batteries and performed a check on the right side first. They saw a doctor and phone screen. After, they performed a check on the left side and they saw ¿on and ok¿ for the left side. The patient outcome was not provided. Further follow-up is being conducted to obtain this information. If additional information is received, a follow-up report will be submitted.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4080014
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #75 on: October 13, 2014, 02:03:08 AM »

Model Number 7426
Event Type Injury
Event Description
It was reported that there was a loss of therapeutic effect. It was noted that the patient was new to the healthcare provider (hcp) and the first time they saw this patient was the day of the report and didn¿t have any medical history on the patient. The patient had been falling, up to 20 times a day. The patient may have many falls during a day or none at all. The patient had a gradual change in benefit over the past year. The right implantable neurostimulator (ins) was dead. The left ins was at 3. 71volts but there was concern about the therapy impedance on the left ins. There was concern about lower therapy impedance of 486ohms with a current of 57. The patient was programmed to 1- 2-c+, 3. 3 volts, rate of 60 microseconds, frequency of 130 hertz, with 24 hours, 7 days a week use. The longevity was 47 months. Electrode impedances measured were in normal range. The patient symptoms were the same on both sides at the time of the report, even though one ins was dead and the left side of the body symptoms may have been a little bit worse than the right side. The hcp didn¿t think the patient was getting much benefit from the ins that was on. No x-rays were taken and no programming was tried yet. The patient was last seen in (b)(6) 2014 by another hcp, but the ins was not checked at that time. The ins status was rechecked, and was ¿showing ok and at 3. 71volts and ok. ¿ electrode impedances were done again and were very similar to the values seen earlier, and the therapy impedance was 488 ohms, current of 57. The ins was on coming into the session, total hours used was 9283 hours. This had been a gradual change for the patient over the past year. The patient had falls in the past and has hit his head. It was noted that the hcp would do an ins replacement. It was later noted that the hcp programed the patient to 3. 6 volts on the left ins and the patient had a little improvement in finger dexterity. There were no concerns about the end of service (eos) or ins longevity on the ins that was dead. It was determined that the cause of the event was device related. The right ins was not explanted or replaced at the time of the report. It was unknown how the patient was doing. They were awaiting a neurosurgery appointment to replace the ins, at the time of the report, but it was unclear which ins was going to be replaced. Information regarding patient outcome and status on device explant has been requested which was not reported regarding this event. If additional information is received, a follow-up report will be sent. Please refer to manufacturer report # 3004209178-2014-16728. It was unclear which ins was being replaced.

Manufacturer Narrative
Concomitant products: product id 7426, serial # (b)(6), implanted: (b)(6) 2011, product type implantable neurostimulator; product id 3387-40, lot # j0527324v, implanted: (b)(6) 2006, product type lead; product id 748251, serial # (b)(6), implanted: (b)(6) 2006, product type extension; product id neu_unknown_ext, lot # unknown, product type extension; product id neu_unknown_lead, lot # unknown, product type lead. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4078161
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #76 on: October 13, 2014, 03:19:05 AM »

Model Number 37602
Event Type Malfunction
Event Description
It was reported that one of the patient¿s falls that had occurred about a year prior to the date of this report had most likely caused a short in the lead in the patient¿s brain. The healthcare professional did not think that the patient was a good candidate for surgery as the patient took plavax which was a blood thinner. The patient took this due to a heart attack that the patient had had 2 years prior to the date of this report. The patient had already had a battery replacement once but it had only lasted 3 months and the rechargeable system was recommended to the patient. Reference manufacturer¿s report number: 3004209178-2014-00651.

Manufacturer Narrative
Concomitant products: product id 37602, serial # (b)(6), implanted: (b)(6) 2012, explanted: (b)(6) 2013, product type implantable neurostimulator; product id 7482a51, serial # (b)(4), implanted: (b)(6) 2010, product type extension; product id 3387s-40, lot # v499735, implanted: (b)(6) 2010, product type lead; product id 37602, serial # (b)(4), implanted: (b)(6) 2013, product type implantable neurostimulator; product id 3387s-40, lot # v020086, implanted: (b)(6) 2007, product type lead; product id 7482a51, serial # (b)(4), implanted: (b)(6) 2007, product type extension; product id 37602, serial # (b)(4), implanted: (b)(6) 2012, explanted: (b)(6) 2013, product type implantable neurostimulator. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4050976
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #77 on: November 07, 2014, 10:58:06 PM »

Model Number 7426
Event Type Injury
Event Description
It was reported that the patient was seen last in (b)(6) 2013 and they would be seen by their clinician 2 days following report. The electrode impedances were measured and they were all fine and normal. The ins that was implanted on their right chest had a slight tingling sensation to their arm. Additional information received reported the patient still had concerns with their device or therapy, but they were working with their manufacturer representative or doctor. It was stated they had an appointment scheduled for (b)(6) 2014. Additional information received reported there was a 50% or greater symptom reduction. It was later reported that the patient was scheduled to have both of the implantable neurostimulators (inss) replaced on (b)(6) 2014. The patient was seen on (b)(6) 2014 for adjustment and reprogramming. The right stimulator was interrogated and the patient had mild paresthesia in the left arm. Right side settings were at 4. 1v, 90usec, 185hz, and 3+2- with minimal tremor in the left arm. The patient was seen on (b)(6) 2014 and there was no evidence of discontinuity of the leads of extensions. The patient¿s was also seen on (b)(6) 2014 for a device adjustment, weakness and gait difficulty. The weakness onset had been gradual and occurring in a persistent pattern for 4 months prior to the appointment, but had first occurred more than 40 years of age. Weakness had been characterized as difficulty arising from a chair, climbing stairs, walking, and increased weakness following exertion and at the end of the day. Weakness was located in the legs and on the right side only. It was associated with numbness. Patient had falls, difficulty ambulating and a gait disturbance. Onset for gait issues had been gradual starting about 8 months prior to the appointment, but was worsening. The patient had slow speech, romberg sign positive, impaired toes, and was tandem walking impaired which were noted during a physical exam. The patient was still experiencing frequent shocks with the battery with fluctuating responses of the battery. The patient was going to be scheduled for a neurosurgery appointment for a battery change. For numbness and tingling, low back pain and right leg weakness an electromyography (emg) and computerized tomography (ct) were going to be performed. Reference manufacturer¿s report number: 3004209178-2014-12463 for the patient¿s other implanted system and the shocking issues that were isolated to that system.

Manufacturer Narrative
Concomitant medical products: product id: 7426, serial# (b)(4), implanted: (b)(6) 2010, product type: implantable neurostimulator. Product id: 7482a51, serial# (b)(4), implanted: (b)(6) 2010, product type: extension. Product id: 3387s-40, lot# v479548, implanted: (b)(6) 2010, product type: lead. Product id: 7482a51, serial# (b)(4), implanted: (b)(6) 2009, product type: extension. Product id: 3387s-40, lot# v368056, implanted: (b)(6) 2009, product type: lead. Product id: 3387s-40, lot# v479548, implanted: (b)(6) 2010, product type: lead. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4210587
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #78 on: November 08, 2014, 03:31:13 AM »

Model Number 37601
Event Type Injury
Event Description
It was reported that the patient¿s programmer was making increases in voltage too quickly and not working appropriately. The patient¿s wife reported that he was having difficulties for a couple of weeks, but thought it was user error. The wife tried it on the day of the report and found that when she made an adjustment it moved to the maximum setting instead of by 0. 1. The nurse practitioner (np) was working with the wife and it seemed as though she was using the programmer appropriately. The np stated that the right lead was the issue and it jumped from 1. 6 volts to 2. 2 volts when trying to increase by 0. 1 volt. This jumping had caused the patient to fall several times and he felt a heavy jolt. The wife stated that the ¿stimulation was spiking, and they made sure that it was not user error on their part, and once she was decreasing the setting, his one setting spiked from 0. 7 to 2. 2 which sent him into a convulsion when this spike occurred and they could not figure out why it was occurring, but it dropped him to the floor about five minutes ago, but it had been going on for a couple of weeks. ¿ the wife stated that this only happened when making adjustments with the programmer and she did not think any of the buttons were stuck down. The wife noted ¿it was the setting on the other side that was spiking, the right side remained consistent. ¿ upon return of the patient programmer analysis found the antenna jack broken. The top case assembly was scratched and the bottom case assembly was damaged. C300. The patient outcome was not reported or if a new programmer resolved the issue, so additional information was requested. If additional information is received a supplemental report will be sent.

Manufacturer Narrative
Concomitant products: product id 37642, serial # (b)(4), product type programmer, patient; product id 3389s-40, lot # va02y9y, implanted: (b)(6) 2012, product type lead; product id 3708660, serial # (b)(4), implanted: (b)(6) 2012, product type extension; product id 3708660, serial # (b)(4), implanted: (b)(6) 2012, product type extension. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4166830
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #79 on: January 10, 2015, 11:26:45 AM »

Model Number 37603
Event Type Injury
Event Description
It was reported the patient had been falling. They fell and broke their collar bone and had lacerations on his head. The patient had fallen 13 times since april. It was stated they had one fall in rehab after implant and then started to have more falls after rehab. Impedances were noted as fine. They had turned the deep brain stimulator (dbs) off to see if the patient continued to fall. No interventions or outcome were reported regarding this event. Further follow-up is being conducted to obtain this information. If additional information is received, a follow-up report will be sent. Please see manufacturer report #3004209178-2014-22314 for information on the patient's concomitant system.

Manufacturer Narrative
Concomitant medical products: product id: 37603, serial# (b)(4), implanted: (b)(6) 2014, product type: implantable neurostimulator. Product id: 3387s-40, lot# va0ery5, implanted: (b)(6) 2014, product type: lead. Product id: 37642, serial# (b)(4), product type: programmer, patient. Product id: 3708660, serial# (b)(4), implanted: (b)(6) 2014, product type: extension. Product id: 3708660, serial# (b)(4), implanted: (b)(6) 2014, product type: extension. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4287020
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #80 on: February 07, 2015, 06:45:23 AM »

Model Number 37602
Event Type Injury
Manufacturer Narrative
Concomitant medical products: product id: 37602, serial# (b)(4), implanted: (b)(6) 2013, product type: implantable neurostimulator. Product id: 3389s-40, lot# v091844, implanted: (b)(6) 2008, product type: lead. Product id: 7482a51, serial# (b)(4), implanted: (b)(6) 2008, product type: extension. Product id: 7482a51, serial# (b)(4), implanted: (b)(6) 2008, product type: extension. Product id: 3389s-40, lot# v032588, implanted: (b)(6) 2008, product type: lead product id: 7438, serial# (b)(4), implanted: (b)(6) 2008, product type: programmer, patient. (b)(4).

Event Description
It was reported that the patient had been admitted to the hospital for frequent falls. It had started on sunday prior to the date of this report and the patient had fallen three times on that day. The patient¿s behaviors were not like they used to be, the patient was hallucinating, agitated and combative. They wanted a manufacturing representative to come assess the implantable neurostimulator (ins). The patient had parkinson¿s disease. No outcome was provided. Further follow-up is being conducted to obtain this information. If additional information is received a supplemental report will be submitted.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4434518
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #81 on: February 07, 2015, 06:46:01 AM »

Model Number 37612
Event Type Injury
Manufacturer Narrative
Concomitant medical products: product id 3387-40, lot# j0421698v, implanted: (b)(6) 2004, product type: lead; product id 748251, serial# (b)(4), implanted: (b)(6) 2004, product type: extension; product id 748251, serial# (b)(4), implanted: (b)(6) 2004, product type: extension; product id 3387-40, lot# j0421511v, implanted: (b)(6) 2004, product type: lead; product id 37642, serial# (b)(4), product type: programmer, patient; product id 37752, serial# (b)(4), product type: recharger. (b)(4).

Event Description
It was reported that there was a shocking or jolting sensation. The patient¿s device had gotten disconnected when he had fallen and he would get a buzz or shock in his chest when he would adjust it. It was unknown when the fall was because he had fallen several times. It had been sometime in-between the implant and his replacement surgery. No outcome was provided. Further follow-up is being conducted to obtain this information. If additional information is received a supplemental report will be submitted.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4409062
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #82 on: February 07, 2015, 09:14:21 AM »

Model Number 37601
Event Type Malfunction
Event Description
It was reported that the patient had the device implanted on (b)(6) 2014 for parkinson¿s and the patient believed the right side was disconnected. The patient had a return of tremors on the left side, he could not walk and kept falling. It was noted that when the patient was taking a shower he had rubbed his head kind of hard and scratched it a little and when he pressed down on his head he could feel tingling in his feet and legs on the left side. This had all begun 3-4 days prior to the date of this report. The patient had called and scheduled an appointment for (b)(6) 2014. No intervention or outcome was provided. Further follow-up is being conducted to obtain this information. If additional information is received a supplemental report will be submitted.

Manufacturer Narrative
Concomitant medical products: product id 3389s-40, lot# va0d7cw, implanted: (b)(6) 2014, product type: lead; product id 3389s-40, lot# va0fgjj, implanted: (b)(6) 2014, product type: lead; product id 3708660, serial# (b)(4), implanted: (b)(6) 2014, product type: extension; product id 3708660, serial# (b)(4), implanted: (b)(6) 2014, product type: extension. (b)(4). (b)(6).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4388049
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #83 on: February 07, 2015, 11:19:58 AM »

Model Number 37602
Event Type Injury
Event Description
It was reported that the device had worked wonderful for the patient¿s upper extremities, but they had dystonia in their left foot that caused their foot to be at almost a 90 degree angle. It had helped with all other symptoms besides the patient¿s foot. This made it impossible for the patient to walk and they kept falling. The patient had this problem before implant. The patient¿s healthcare provider (hcp) instructed them to turn the device off. The hcp thought they could help fix the issues, but wanted the patient to turn stimulation off for a week to week and a half, so they could start over. The patient later reported that he turned stimulation off, but program c was still on. About a month later the patient still had concerns regarding his device or therapy, but was working with their hcp or a manufacturer representative. The patient had an appointment on (b)(6) 2014. The patient knew what a, b, and c were, but wanted to know what they control. The patient wrote that a felt like it was in their legs, b in their voice and they were stuttering, and they did not know what c did. The patient stated they used all three. Six months later the patient reported the patient programmer screen showed the ¿check the implantable neurostimulator (ins)¿ screen. There was a problem with the programmer and the programmer worked intermittently. When the patient checked the ins they got the stimulation turned off screen. The patient tried to turn stimulation back on, but they could not make a connection. The patient saw stimulation off and a flashing triangle with the battery on the top row of the programmer and 3. 108 below that. Usually the patient could feel a small voltage going through their body when they move to program b or c, but they did not feel anything at the time of this report. The patient¿s wife stated that it had been worrying the patient a good bit because they had to use the ins more than ever. The patient had symptoms of slurred speech. The patient had turned the ins off for a short period of time and they have never had any trouble with it so they assumed it was the batteries. The reporter stated the patient was beginning to get it a little on the right side. The patient had dystonia real bad and ¿afib¿ that caused them to fall almost every day. When the patient¿s hcp ¿hooked up the final wires¿ the patient was walking again, but their left foot was turning in. The reporter stated the hcp told them that normally the patients get better and they had never seen one get so bad. The patient¿s right foot turned towards the middle and they had a brace on it. The patient did use a walker to walk. The patient¿s foot started turning in about three weeks after surgery. For the past six months the patient was starting to get dizzy when they stood up and tried to take a step. The patient had fallen once close to thanksgiving and then again about six weeks prior to this report. After the first fall the patient had a concussion and after the second fall the patient was taken by ambulance to the hospital because they thought their hip may be broken. The patient had been to the orthopedic hcp twice since then. No interventions or outcome were reported regarding this event. Further follow-up is being conducted to obtain this information. If additional information is received, a follow up report will be sent.

Manufacturer Narrative
Concomitant medical products: product id 37642, serial# (b)(4), product type: programmer, patient; product id 3 387s-40, lot# va09grv, implanted: (b)(6) 2013, product type: lead; product id 748351, serial# (b)(4), implanted: (b)(6) 2013, product type: extension; product id 37642, serial# (b)(4), product type: programmer, patient. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4386252
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #84 on: March 07, 2015, 04:40:39 AM »

Model Number 7426
Event Type Injury
Event Description
It was reported the patient had been having a lot of problems with falling and they were in a wheelchair. A patient injury was reported. The reporter stated they thought the patient fell 37 times before they had surgery for their back in (b)(6) 2010. The patient continued to fall since the surgery in 2010. The reporter stated they read something about deep brain stimulation and heard there was a recall that covered a lead so they wondered if the patient¿s implantable neurostimulator (ins) that was removed was part of the recall. The patient¿s ins was replaced on (b)(4) 2014. No outcome was reported regarding this event. Further follow-up is being conducted to obtain this information. If additional information is received, the event will be updated.

Manufacturer Narrative
Concomitant medical products: product id: 3387s-40, lot# v165128, implanted: (b)(4) 2008, product type: lead. Product id: 7482a40, serial# (b)(4), product type: extension. Product id: 3387s-40, lot# v165128, implanted: (b)(6) 2008, product type: lead. Product id: neu_unknown_prog, product type: programmer, physician. (b)(4).

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4484557
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #85 on: April 19, 2015, 01:22:11 AM »

Model Number 37603
Event Type Injury
Event Description
It was reported the patient had a total of 3 surgeries for the deep brain stimulator (dbs) system. They had done well with the first two surgeries but the third surgery ¿totally did him in. ¿ he had lost muscle strength, had seizures, big falls, and towards the end of his life when he was in the hospital he could not control moving his right foot. The patient was told they had parkinson¿s disease but originally the patient was not diagnosed with that and he just had essential tremors pre-dbs surgeries. No further follow-up is being done on this event as the patient died and it was stated to be unrelated to the device or therapy. The death occurred following the event and therefore, unable to follow-up for patient outcome. Please see manufacturer report #3004209178-2015-04324 for information on the patient's concomitant system.

Manufacturer Narrative
Concomitant medical products: product id 37603, serial# (b)(4), implanted: (b)(6) 2011, product type: implantable neurostimulator; product id 3387s-40, lot# v556377, implanted: (b)(6) 2011, product type: lead; product id 37642, serial# (b)(4), product type: programmer, patient; product id 37085-40, serial# (b)(4), implanted: (b)(6) 2011, product type: extension; product id 3387s-40, lot# v603771, implanted: (b)(6) 2011, product type: lead; product id 37085-40, serial# (b)(4), implanted: (b)(6) 2011, product type: extension. (b)(4).

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4577570
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #86 on: April 19, 2015, 01:23:07 AM »

Model Number 37603
Event Date 01/08/2015
Event Type Injury
Manufacturer Narrative
Product id 3387s-40, lot# va0nfax, implanted: 2014 (b)(6); product type lead product id 37642, serial# (b)(4); product type programmer, patient product id 3708660, serial# (b)(4), implanted: 2014 (b)(6); product type extension. (b)(4).

Event Description
It was reported, the patient had developed a large left subdural hematoma with 6 mm midline shift to the right. Observation and repeated scans were done as an intervention. A ct scan without contrast showed large left convexity mixed age subdural hematoma effacing left hemispheric sulci and resulting in 6 mm midline shift to the right. The patient was scanned again in (b)(6) and showed a chronic left subdural hematoma that had continued mixed density suggesting there had been further hemorrhage. The volume of the subdural hematoma was slightly larger and the left to right shift measured 7 mm. The patient was scanned again a few days later and showed the mixed density subdural hematoma collection was not significantly changed. The patient underwent an elective burr hole evacuation a couple days after the last ct scan. The evacuation did not resolve the subdural hematoma. The patient also began to have weakness with falls after the subdural hematoma was first noted. The evacuation date was moved up due to the falls. The event required medical or surgical intervention and resulted in in-patient hospitalization and prolonged existing hospitalization. The etiology was the lead/extension tract and was unlikely related to the device or therapy and possibly related to the implant procedure. The event was ongoing with no further action needed. If additional information is received a follow-up report will be sent.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4638254
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #87 on: May 09, 2015, 02:04:45 AM »

Model Number 37602
Event Type Injury
Event Description
It was reported that the patient had started going downhill in (b)(6) 2014. The patient had started falling a lot in (b)(6) 2014 which was why the patient had to be put into a nursing home on (b)(6) 2015. It was noted that after one of the patient¿s falls onto the floor the patient was placed in a hospital but it was unknown when the patient was admitted or when the fall was. The patient was released from the hospital on (b)(6) 2015 and was then admitted to the nursing care facility. The patient had fallen through the shower door, by the shower and a couple of times by the bed. The healthcare professional had done x-rays and everything was fine. The patient had very low blood pressure and took medication to bring the blood pressure up. It was noted that there were also times where he got up right away and he should not be doing that. The patient was in a wheelchair and does not walk with his walker. Starting in (b)(6) 2014 the patient had also been having trouble swallowing anything. The patient coughed a lot when he ate and even when he drank water. One day the patient had coughed really badly for a quite a while. The patient¿s wife pureed the food for patient and added thickening for liquids. On (b)(6) 2015 they had done a swallowing/aspiration test and it was pretty bad. The patient might get a g-tube for feeding purposes. The patient¿s last battery had lasted from 2005 to 2011 and they thought this battery would last for 5 years. The patient¿s family thought the battery was getting low because the patient was having neurological and swallowing issues which were symptoms that the patient usually had when their battery was getting low. The family wanted the implantable neurostimulator (ins) checked before placing the g-tube. The healthcare professional was unsure if the patient was getting benefit from the deep brain stimulator. No intervention or outcome was provided. Further follow-up is being conducted to obtain this information. If additional information is received a supplemental report will be submitted. Reference manufacturer¿s report number: 3004209178-2015-07114.

Manufacturer Narrative
Concomitant medical products: product id: 37602, serial# (b)(4), implanted: (b)(6) 2011, product type: implantable neurostimulator. Product id: 3389s-40, lot# v010293, implanted: (b)(6) 2006, product type: lead. Product id: 748251, serial# (b)(4), implanted: (b)(6) 2006, product type: extension. Product id: 748295, serial# (b)(4), implanted: (b)(6) 2006, product type: extension. (b)(4).

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4684090
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #88 on: May 09, 2015, 02:05:31 AM »

Model Number 7426
Event Date 09/15/2013
Event Type Injury
Manufacturer Narrative
(b)(4).

Event Description
It was reported that there was an increase of frequency of falls. The increase of falls occurred after the adjustment of stimulation parameters. The adjustment included a decrease of frequency and augmentation of voltage. The patient experienced hospitalization or prolongation of existing hospitalization. The event was possible or certainly related to the stimulation and medication. The event resolved completely.

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4717965
Logged
dennis100
Moderators
Hero Member
*****

Karma: +24/-0
Offline Offline

Posts: 62863


« Reply #89 on: June 08, 2015, 05:55:37 AM »

Model Number 37603
Event Type Injury
Event Description
It was reported the patient had surgery last (b)(6) and ¿it had not worked since the implant. ¿ it was noted the patient would ¿pick up speed when they would talk. ¿ it was stated since the surgery the patient had fallen down 6-7 times. It was noted the patient had speech impairment and it was hard to understand them. It was later reported that there was a poor communication screen but they were eventually able to conform with both the left and right sides on and ok. Stimulation was turned off and back on. Since implant the patient had been falling down and losing her balance much more than before the implant. The patient¿s balance and falls had gotten worse after implant and the patient was unsure if it was safe to walk and she used a walker. The patient had had a major fall about 2 months after implant,(b)(6) or (b)(6), she had fallen on the carpet while she was hanging clothes. It was noted that the patient went to rehab after the fall. Patient wanted to discuss turning therapy off to see if it would improve her falls/balance. Patient inquired about lead placement and how it effects stimulation. The patient was not sure if the deep brain stimulator was causing her falls but felt that it was either the deep brain stimulator or disease progression. It was later reported that since implant the patient had balance issues. The patient had never had this before implant and had fallen multiple times. The patient was told that it was due to her recent surgery. The patient was still falling and had broken her hand from a fall. Reference manufacturer¿s report number: 3004209178-2015-08800.

Manufacturer Narrative
Concomitant medical products: product id: 37603, serial# (b)(4), implanted: (b)(6) 2012, product type: implantable neurostimulator. Product id: 3387s-40, lot# va01zjk, implanted: (b)(6) 2012, product type: lead. Product id: 37642, serial# (b)(4), product type: programmer, patient. Product id: 3708660, serial# (b)(4) implanted: (b)(6) 2012, product type: extension. Product id: 3387s-40, lot# va01zjk, implanted: (b)(6) 2012, product type: lead. Product id: 3708660, serial# (b)(4), implanted: (b)(6) 2012, product type: extension. Product id 37603 lot# serial#(b)(4), implanted: (b)(6) 2012, product type: implantable neurostimulator. (b)(4).

https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfmaude/detail.cfm?mdrfoi__id=4761849